Phase noise can make or break a synthesizer design, but not everyone has a dedicated phase noise analyzer on the bench. Here's my practical take on what a spectrum analyzer can and can't do.
The SA method: carrier at center, narrow RBW, measure noise power in a 1 Hz-normalized bandwidth at the offset of interest, subtract carrier power. Simple on paper. In practice:
1. The SA's own LO phase noise sets the floor. If your DUT is within ~10 dB of the SA's own noise, you're measuring the analyzer, not the DUT. Check the SA datasheet first — this kills more measurements than any other mistake.
2. AM noise leaks in. The SA can't separate AM from PM noise. If your source has significant AM noise (many PLLs do at close-in offsets), your "phase noise" number is pessimistic. A quick sanity check: AM-detector measurement or compare against a known-clean reference.
3. Close-in offsets (<1 kHz) are where SAs struggle most — 1/f noise of the SA's own LO dominates.
When to rent or borrow a dedicated test set: characterizing OCXOs, low-noise DROs, or any time you need close-in data you can actually trust. Cross-correlation instruments buy you 10–20 dB of extra sensitivity — worth it for a week-long rental when a design decision hangs on the number.
Practical tips that helped me:
- Always measure with the DUT at the same drive level you'll use in the system; PN is level-dependent.
- Warm up both instruments for at least 30 minutes.
- Log the SA's own noise floor measurement alongside every DUT plot — future-you will thank present-you.
What's the lowest phase noise you've credibly measured with just a spectrum analyzer, and how did you convince yourself the number was real?
The SA method: carrier at center, narrow RBW, measure noise power in a 1 Hz-normalized bandwidth at the offset of interest, subtract carrier power. Simple on paper. In practice:
1. The SA's own LO phase noise sets the floor. If your DUT is within ~10 dB of the SA's own noise, you're measuring the analyzer, not the DUT. Check the SA datasheet first — this kills more measurements than any other mistake.
2. AM noise leaks in. The SA can't separate AM from PM noise. If your source has significant AM noise (many PLLs do at close-in offsets), your "phase noise" number is pessimistic. A quick sanity check: AM-detector measurement or compare against a known-clean reference.
3. Close-in offsets (<1 kHz) are where SAs struggle most — 1/f noise of the SA's own LO dominates.
When to rent or borrow a dedicated test set: characterizing OCXOs, low-noise DROs, or any time you need close-in data you can actually trust. Cross-correlation instruments buy you 10–20 dB of extra sensitivity — worth it for a week-long rental when a design decision hangs on the number.
Practical tips that helped me:
- Always measure with the DUT at the same drive level you'll use in the system; PN is level-dependent.
- Warm up both instruments for at least 30 minutes.
- Log the SA's own noise floor measurement alongside every DUT plot — future-you will thank present-you.
What's the lowest phase noise you've credibly measured with just a spectrum analyzer, and how did you convince yourself the number was real?