Not everyone has a spectrum analyzer on the bench, but most of us have a scope with FFT. With some care it works surprisingly well for EMI pre-compliance. A few things I learned the hard way:
1. RBW is everything
Narrow RBW = lower noise floor but slower sweep. For conducted emissions (150 kHz-30 MHz), I use 9 kHz RBW to roughly match CISPR receivers. For radiated pre-scans, 120 kHz. Wide RBW hides narrowband spurs.
2. Window function matters more than people think
Use flattop when you care about amplitude accuracy (comparing against a limit line), Hanning for general spur hunting. Rectangular looks sharper but its sidelobes will lie to you.
3. Probe like you mean it
A near-field H-probe with 20-30 dB preamp gets you surprisingly close to chamber results for relative comparisons. Keep the probe ground lead short — better yet, use a ground spring. On cables, a common-mode current clamp tells you more than any near-field probe.
4. Know the limits
Scope FFT dynamic range is ~50-60 dB at best, nowhere near a real receiver. It is great for "did my fix help" A/B comparisons, not for certifying anything.
My workflow: near-field scan to find the hot spots, current clamp on cables to rank them, then fix, re-scan, compare. Quick and honest.
What's your go-to pre-compliance setup? Anyone correlating scope-FFT numbers against chamber data with a consistent offset?
1. RBW is everything
Narrow RBW = lower noise floor but slower sweep. For conducted emissions (150 kHz-30 MHz), I use 9 kHz RBW to roughly match CISPR receivers. For radiated pre-scans, 120 kHz. Wide RBW hides narrowband spurs.
2. Window function matters more than people think
Use flattop when you care about amplitude accuracy (comparing against a limit line), Hanning for general spur hunting. Rectangular looks sharper but its sidelobes will lie to you.
3. Probe like you mean it
A near-field H-probe with 20-30 dB preamp gets you surprisingly close to chamber results for relative comparisons. Keep the probe ground lead short — better yet, use a ground spring. On cables, a common-mode current clamp tells you more than any near-field probe.
4. Know the limits
Scope FFT dynamic range is ~50-60 dB at best, nowhere near a real receiver. It is great for "did my fix help" A/B comparisons, not for certifying anything.
My workflow: near-field scan to find the hot spots, current clamp on cables to rank them, then fix, re-scan, compare. Quick and honest.
What's your go-to pre-compliance setup? Anyone correlating scope-FFT numbers against chamber data with a consistent offset?