Displayed average noise level looks like a single number on a datasheet, but on the bench it moves with almost every setting you touch. When I need to measure a low-level spur that sits close to the noise floor, I work through a fixed checklist before trusting any reading.
First, set the resolution bandwidth deliberately. Narrower RBW lowers the displayed noise, but it also slows the sweep and can hide a spur that drifts. I start wide to find the spur, then narrow the RBW in steps and watch whether the spur level stays stable while the noise floor drops. If the spur moves with RBW, it is probably not a real continuous signal.
Second, check the attenuator and preamp combination. Too much input attenuation buries the spur; too little risks compression and creates false spurs of its own. I step the attenuation and confirm the spur level changes by the expected amount. A spur that does not track attenuation is usually generated inside the analyzer.
Third, use averaging correctly. Trace averaging reduces displayed noise variance, but it does not improve DANL the way a narrower RBW does, and it can smooth away an intermittent spur. For drifting or bursty signals, I compare a max-hold trace against an averaged trace rather than trusting either one alone.
Fourth, verify with a different span and sweep time. A spur that only appears at one span or one sweep setting deserves suspicion: check for image responses, LO feedthrough, and residual responses by shifting the center frequency slightly. A real spur stays put; an analyzer artifact moves or vanishes.
Finally, record the full setup: RBW, VBW, attenuation, preamp state, and detector type. Two engineers measuring the same spur with different settings will get different numbers, and without the setup the numbers are not comparable.
What is your go-to sanity check when a spur sits within a few dB of the noise floor? Do you trust max-hold, averaging, or a narrower RBW first?
First, set the resolution bandwidth deliberately. Narrower RBW lowers the displayed noise, but it also slows the sweep and can hide a spur that drifts. I start wide to find the spur, then narrow the RBW in steps and watch whether the spur level stays stable while the noise floor drops. If the spur moves with RBW, it is probably not a real continuous signal.
Second, check the attenuator and preamp combination. Too much input attenuation buries the spur; too little risks compression and creates false spurs of its own. I step the attenuation and confirm the spur level changes by the expected amount. A spur that does not track attenuation is usually generated inside the analyzer.
Third, use averaging correctly. Trace averaging reduces displayed noise variance, but it does not improve DANL the way a narrower RBW does, and it can smooth away an intermittent spur. For drifting or bursty signals, I compare a max-hold trace against an averaged trace rather than trusting either one alone.
Fourth, verify with a different span and sweep time. A spur that only appears at one span or one sweep setting deserves suspicion: check for image responses, LO feedthrough, and residual responses by shifting the center frequency slightly. A real spur stays put; an analyzer artifact moves or vanishes.
Finally, record the full setup: RBW, VBW, attenuation, preamp state, and detector type. Two engineers measuring the same spur with different settings will get different numbers, and without the setup the numbers are not comparable.
What is your go-to sanity check when a spur sits within a few dB of the noise floor? Do you trust max-hold, averaging, or a narrower RBW first?