Short PA measurements look clean. Then you run a two-hour EVM test and the numbers slowly walk away from you. Gain drifts down a fraction of a dB, EVM creeps up, and by the end of the session your morning data no longer matches your afternoon data. This is thermal drift, and it is one of the most common reasons bench results fail to reproduce.
Here is how I track it down:
First, separate device drift from setup drift. Put a directional coupler and a power sensor on the PA output and log output power continuously during a long CW run. If the drift follows the heatsink temperature with a lag of minutes, it is the device and its matching network heating up. If it jumps when the lab air conditioning cycles, it is the setup — cables, attenuators, or the signal generator itself.
Second, watch the bias current, not just the gain. As the junction heats, a fixed gate bias lets the drain current walk. Log Idq alongside temperature: a PA that starts at 100 mA and ends at 130 mA is telling you its bias network has no thermal compensation. A simple temperature-compensating bias circuit, or just re-biasing at operating temperature, often recovers most of the drift.
Third, measure EVM as a function of time, not as a single number. I run the modulated signal for 30 minutes and log EVM every minute. The pattern matters: a slow monotonic rise points to thermal settling; a sawtooth points to the heatsink fan cycling; a sudden step points to a protection circuit kicking in.
Fourth, control the thermal environment before blaming the device. A PA bolted to a bench with no heatsink compound, or one sitting in still air under a test fixture lid, will drift far more than the datasheet suggests. Torque the mounting screws evenly, use a proper thermal interface, and give the device 15 minutes of warm-up at the test power before you start logging data.
The mistake I see most often is calibrating the setup cold and then measuring hot. Calibrate at operating temperature, log temperature alongside every measurement, and the drift stops being a mystery.
How do you handle thermal drift in your long PA test sessions — warm-up soak, active cooling, or bias compensation?
Here is how I track it down:
First, separate device drift from setup drift. Put a directional coupler and a power sensor on the PA output and log output power continuously during a long CW run. If the drift follows the heatsink temperature with a lag of minutes, it is the device and its matching network heating up. If it jumps when the lab air conditioning cycles, it is the setup — cables, attenuators, or the signal generator itself.
Second, watch the bias current, not just the gain. As the junction heats, a fixed gate bias lets the drain current walk. Log Idq alongside temperature: a PA that starts at 100 mA and ends at 130 mA is telling you its bias network has no thermal compensation. A simple temperature-compensating bias circuit, or just re-biasing at operating temperature, often recovers most of the drift.
Third, measure EVM as a function of time, not as a single number. I run the modulated signal for 30 minutes and log EVM every minute. The pattern matters: a slow monotonic rise points to thermal settling; a sawtooth points to the heatsink fan cycling; a sudden step points to a protection circuit kicking in.
Fourth, control the thermal environment before blaming the device. A PA bolted to a bench with no heatsink compound, or one sitting in still air under a test fixture lid, will drift far more than the datasheet suggests. Torque the mounting screws evenly, use a proper thermal interface, and give the device 15 minutes of warm-up at the test power before you start logging data.
The mistake I see most often is calibrating the setup cold and then measuring hot. Calibrate at operating temperature, log temperature alongside every measurement, and the drift stops being a mystery.
How do you handle thermal drift in your long PA test sessions — warm-up soak, active cooling, or bias compensation?