Started by LunaWei · 2026-11-14 01:53 UTC · 43 views · 1 replies
L Admin
[Technical article / FAQ, rfforum Editorial Team]

A campaign that slowly walks 0.1–0.2 dB after "nothing changed on the DUT" sometimes tracks the torque wrench, not the product. An uncalibrated or worn wrench under-torques one day and mushrooms the contact the next; the reference plane wanders while the traveler still says "hand plus a click."

1. What the wrench owns
- Under-torque leaves a soft contact that moves with flex and temperature; over-torque mushrooms the female and permanently raises reflection and loss.
- A wrench that clicks early or late by even a small fraction of the specified torque invents a residual that looks like DUT IL drift across mates.
- Swapping wrenches mid-campaign without a thru check after the swap is the classic unlogged change.

2. Freeze list before you blame the DUT
- Wrench PN, last calibration date, and the torque value written on the traveler
- A thru re-measure after cleaning + correct torque with a second known-good wrench
- Whether the 0.15 dB shift appeared after a wrench change, a drop of the tool, or a new operator
- Adapter cycle count: a good wrench on a ruined adapter still fails

3. Habit
Never close a sub-0.2 dB IL drift without wrench ID + cal date + thru check. Archive torque value and wrench PN next to every acceptance IL. If a calibrated wrench and clean adapter kill the delta, the DUT was never the patient.

4. Practical five-minute gate
- Verify wrench cal sticker; replace if overdue or dropped
- Clean ferrules; torque to spec; re-measure thru
- If the delta is gone, stop the DUT ECO and retire or recalibrate the bad wrench

Discussion: on your last 0.15 dB mystery, did the fix land in the wrench, the adapter, or the DUT — and do you require wrench PN + cal date on every IL traveler?

— rfforum 编辑部 / Editorial Team
L Admin
Addendum — sub-0.2 dB IL traveler: wrench PN, last calibration date, and the torque value, on the same line as the acceptance IL. A campaign that walks 0.15 dB after “nothing changed on the DUT” often tracks the tool or the adapter, not the product.

A cheap gate: clean, torque with a second known-good calibrated wrench, and re-measure a thru. If the delta is gone, retire or recalibrate the first wrench and stop the DUT ECO. A good wrench on a ruined adapter still fails — cycle count belongs next to the wrench ID.

Open: on your last 0.15 dB mystery, did the fix land in the wrench, the adapter, or the DUT — and do you require wrench PN + cal date on every IL traveler?

— rfforum 编辑部 / Editorial Team