Conducted emissions testing looks simple on paper — LISN, receiver, sweep — but I keep seeing the same grounding mistakes inflate readings by 10 dB or more, and then people chase a filter problem that does not exist.
1. Bond the LISN properly: the LISN ground plane must be bonded to the chamber ground plane with a low-impedance strap, not a long green wire. A 30 cm wire at 30 MHz is an inductor, and the voltage drop across it shows up directly in your measurement.
2. Keep the EUT ground reference short: the EUT sits on the ground plane with its own bonding strap. If the EUT floats on a wooden bench with a single wire to ground, the common-mode loop area changes and the reading follows the setup, not the product.
3. Mind the cable layout: the mains cable between LISN and EUT should be short, routed flat on the ground plane, and kept away from the measurement cable. A cable draped in the air couples differently every time you move it, which is why repeat measurements disagree.
4. Separate the receiver ground: the measurement receiver should share the same ground plane, but do not daisy-chain grounds through the receiver chassis. One star point, short bonds, and no ground loops through the bench.
5. Sanity-check with a known source: before blaming the EUT, measure a reference load through the same setup. If the reference also reads high, the setup is lying, not the product.
What grounding mistake cost you the most debug time on an emissions failure?
1. Bond the LISN properly: the LISN ground plane must be bonded to the chamber ground plane with a low-impedance strap, not a long green wire. A 30 cm wire at 30 MHz is an inductor, and the voltage drop across it shows up directly in your measurement.
2. Keep the EUT ground reference short: the EUT sits on the ground plane with its own bonding strap. If the EUT floats on a wooden bench with a single wire to ground, the common-mode loop area changes and the reading follows the setup, not the product.
3. Mind the cable layout: the mains cable between LISN and EUT should be short, routed flat on the ground plane, and kept away from the measurement cable. A cable draped in the air couples differently every time you move it, which is why repeat measurements disagree.
4. Separate the receiver ground: the measurement receiver should share the same ground plane, but do not daisy-chain grounds through the receiver chassis. One star point, short bonds, and no ground loops through the bench.
5. Sanity-check with a known source: before blaming the EUT, measure a reference load through the same setup. If the reference also reads high, the setup is lying, not the product.
What grounding mistake cost you the most debug time on an emissions failure?