A noise-figure number that moved 0.2-0.4 dB after the ENR source sat on the bench, or after you moved it from a cart, is often not a DUT change. The ENR table is only valid at the temperature and connector state the cal lab assumed. A warm body, a hot attenuator in front, or a stale table invents a Y-factor walk that follows the 346 more than the amplifier.
1. What the source setup owns
- ENR is a function of the physical temperature of the noise diode and of any pad between the diode and the reference plane. A source left next to a PA, under a lamp, or held in a hand can move several tenths of a dB before the meter looks "settled."
- A table taken at 23 C and used at 30 C, or a table that still includes a pad you removed, is a systematic offset, not a noisy DUT.
- Excess noise from a poorly torqued or dirty 3.5 mm / SMA interface changes the delivered ENR. One remate that "fixes" NF is a connector clue, not an LNA clue.
2. Freeze list before you retune the LNA
- Source model, serial, ENR-table date, and whether the table includes the pad actually in the path
- Case or diode temperature, or at least "time since power-on / time since moved"
- A controlled remate under marked torque, then a second NF on a known attenuator or a second source
- Receiver / meter warm-up and the Y-factor method settings (averaging, SB, cold-source vs hot/cold)
3. Habit
Never freeze "NF drifted 0.3 dB" from a single scan after the noise source was moved or left on a warm bench. Archive table date, pad state, and a temperature or settle note next to every release NF. If a re-zero of the table, a remate, or a 15-minute settle kills the step, the DUT was never the patient.
4. Practical gate
- Leave the source and the first pad dressed in the final position for the settle time in the source manual (or >=15 min if unknown)
- Confirm the ENR table matches the pad stack
- Remate once; if NF jumps, stop the schematic change
Discussion: on your last NF walk that followed the 346 rather than the LNA, did the fix land in a stale ENR table, a warm source, or a remate - and do you require a dressed, settled source before a release number?
— rfforum 编辑部 / Editorial Team
A noise-figure number that moved 0.2-0.4 dB after the ENR source sat on the bench, or after you moved it from a cart, is often not a DUT change. The ENR table is only valid at the temperature and connector state the cal lab assumed. A warm body, a hot attenuator in front, or a stale table invents a Y-factor walk that follows the 346 more than the amplifier.
1. What the source setup owns
- ENR is a function of the physical temperature of the noise diode and of any pad between the diode and the reference plane. A source left next to a PA, under a lamp, or held in a hand can move several tenths of a dB before the meter looks "settled."
- A table taken at 23 C and used at 30 C, or a table that still includes a pad you removed, is a systematic offset, not a noisy DUT.
- Excess noise from a poorly torqued or dirty 3.5 mm / SMA interface changes the delivered ENR. One remate that "fixes" NF is a connector clue, not an LNA clue.
2. Freeze list before you retune the LNA
- Source model, serial, ENR-table date, and whether the table includes the pad actually in the path
- Case or diode temperature, or at least "time since power-on / time since moved"
- A controlled remate under marked torque, then a second NF on a known attenuator or a second source
- Receiver / meter warm-up and the Y-factor method settings (averaging, SB, cold-source vs hot/cold)
3. Habit
Never freeze "NF drifted 0.3 dB" from a single scan after the noise source was moved or left on a warm bench. Archive table date, pad state, and a temperature or settle note next to every release NF. If a re-zero of the table, a remate, or a 15-minute settle kills the step, the DUT was never the patient.
4. Practical gate
- Leave the source and the first pad dressed in the final position for the settle time in the source manual (or >=15 min if unknown)
- Confirm the ENR table matches the pad stack
- Remate once; if NF jumps, stop the schematic change
Discussion: on your last NF walk that followed the 346 rather than the LNA, did the fix land in a stale ENR table, a warm source, or a remate - and do you require a dressed, settled source before a release number?