When the DUT sits a metre or two away from the VNA, the cables stop being invisible. Phase rotates fast with frequency, a small cable movement shifts the measurement, and the reference plane ends up somewhere inside the cable instead of at the DUT. A few things I have learned to do about it.
First, decide where the reference plane should be, then calibrate there. If you calibrate at the VNA ports and just hope the cables are ideal, you are measuring the cables. Do the SOLT or e-cal at the end of the cables, right where the DUT connects. It costs one extra cal step and saves hours of arguing with the data later.
Second, port extension is a delay correction, not a loss correction. It fixes the phase rotation from the electrical length, but it does nothing for cable loss or mismatch. If your cables have real loss, use de-embedding with measured cable S-parameters instead of pretending the extension handles it.
Third, mind cable movement. Flexing a test cable changes its phase, sometimes by degrees at higher frequencies. Once the cal is done, route the cables so nothing tugs on them during the measurement, and avoid re-bending between cal and measurement. If you must move the DUT, re-check with a thru standard afterwards.
Fourth, characterize the adapters. Every adapter between cable and DUT adds its own reflection. A quick one-port check of each adapter against a known load tells you whether the adapter is the thing ruining your return loss, before you blame the DUT.
Fifth, verify with a check standard. After cal, measure a known device — a 20 dB attenuator or a matched load — and compare against its datasheet. If the check standard does not measure right, nothing else you measure that day is trustworthy.
How do you handle long cables in your setup — cal at the DUT, port extension, or full de-embedding?
First, decide where the reference plane should be, then calibrate there. If you calibrate at the VNA ports and just hope the cables are ideal, you are measuring the cables. Do the SOLT or e-cal at the end of the cables, right where the DUT connects. It costs one extra cal step and saves hours of arguing with the data later.
Second, port extension is a delay correction, not a loss correction. It fixes the phase rotation from the electrical length, but it does nothing for cable loss or mismatch. If your cables have real loss, use de-embedding with measured cable S-parameters instead of pretending the extension handles it.
Third, mind cable movement. Flexing a test cable changes its phase, sometimes by degrees at higher frequencies. Once the cal is done, route the cables so nothing tugs on them during the measurement, and avoid re-bending between cal and measurement. If you must move the DUT, re-check with a thru standard afterwards.
Fourth, characterize the adapters. Every adapter between cable and DUT adds its own reflection. A quick one-port check of each adapter against a known load tells you whether the adapter is the thing ruining your return loss, before you blame the DUT.
Fifth, verify with a check standard. After cal, measure a known device — a 20 dB attenuator or a matched load — and compare against its datasheet. If the check standard does not measure right, nothing else you measure that day is trustworthy.
How do you handle long cables in your setup — cal at the DUT, port extension, or full de-embedding?