I would start by separating three quantities before changing the calibration: raw directivity of the VNA plus cable assembly, residual directivity after calibration, and connector repeatability at the moving reference plane. A useful first test is to leave the calibration kit and adapters fixed, calibrate with the cable in its nominal pose, then measure a precision load while moving only through the intended flex envelope. If the ripple amplitude and phase track cable position, the likely issue is cable stability or a mode behind the reference plane. If the trace jumps mainly when the connector is disturbed and then stays offset, suspect torque, pin gap, contamination, or adapter repeatability.
The next diagnostic is a raw-directivity check using a good sliding or offset load before correction, followed by the same check after calibration. Raw directivity tells you whether the hardware front end can support the desired residual performance; calibrated residual directivity tells you whether the standards and calibration process actually removed the reflection. At mmWave, small mechanical changes can look surprisingly large because a short length of damaged cable or a non-repeatable connector creates a vector error comparable to the intended directivity term. Also verify that the cable bend radius, phase stability rating, connector torque, and calibration frequency range are all appropriate for the band.
A practical sequence is: inspect and clean connectors, torque to the specified value, perform a full warm-up, calibrate at the final reference plane, then measure load, short, and open without moving anything. Record magnitude and phase, repeat after disconnecting and reconnecting the same standard, and finally repeat while flexing the cable. Reconnection scatter identifies connector/standard repeatability; flex-only scatter identifies cable and fixture issues; a stable but poor result points to calibration residual, damaged standards, or an inadequate VNA/bridge raw specification. Do not use port extension to hide this problem, because extension changes the reference plane but cannot restore lost directivity or remove multimode reflections behind the calibration plane.
The next diagnostic is a raw-directivity check using a good sliding or offset load before correction, followed by the same check after calibration. Raw directivity tells you whether the hardware front end can support the desired residual performance; calibrated residual directivity tells you whether the standards and calibration process actually removed the reflection. At mmWave, small mechanical changes can look surprisingly large because a short length of damaged cable or a non-repeatable connector creates a vector error comparable to the intended directivity term. Also verify that the cable bend radius, phase stability rating, connector torque, and calibration frequency range are all appropriate for the band.
A practical sequence is: inspect and clean connectors, torque to the specified value, perform a full warm-up, calibrate at the final reference plane, then measure load, short, and open without moving anything. Record magnitude and phase, repeat after disconnecting and reconnecting the same standard, and finally repeat while flexing the cable. Reconnection scatter identifies connector/standard repeatability; flex-only scatter identifies cable and fixture issues; a stable but poor result points to calibration residual, damaged standards, or an inadequate VNA/bridge raw specification. Do not use port extension to hide this problem, because extension changes the reference plane but cannot restore lost directivity or remove multimode reflections behind the calibration plane.