Assumptions first: the E-cal module itself was warmed up and connected with the specified torque, the adapter is electrically stable across the full band, and the VNA calibration kit definition matches the adapter gender and interface. If any of these is unknown, treat the adapter as part of the DUT path rather than assuming it is inside the corrected reference plane.

Step 1 is to repeat a known short/open/load or verification standard at the intended reference plane and compare raw and corrected S11 in both frequency and time domain. A smooth phase slope with little amplitude change usually points to a reference-plane offset or wrong kit delay. A ripple whose period changes with frequency or cable position is more consistent with connector repeatability, cable flexure, or a multimode adapter. Step 2 is to calibrate without the adapter, verify, then add the adapter using a properly defined adapter characterization or removal process. This separates E-cal drift from adapter mismatch and delay.

Do not use port extension as a substitute when the adapter has meaningful return loss or dispersion. Port extension mainly adds ideal phase and sometimes loss; it cannot remove adapter mismatch. If the failure remains after replacing the adapter and restraining the cable, suspect warm-up, E-cal connector wear, or residual directivity, and confirm with a separate verification kit rather than judging only from the calibration residuals displayed by the analyzer.