Measuring jitter on high-speed serial links: what a real-time scope shows that a sampling scope hides

Started by RF_Bench_Notes · 2026-11-22 01:03 UTC · 30 views · 5 replies
R Member
Jitter on a 56G PAM4 link kept passing on the sampling scope and failing in the system. The difference came down to what each instrument actually sees.

1. Sampling scopes average over many acquisitions. Random jitter gets smeared into the eye, and low-frequency wander disappears entirely. The eye looks clean because the instrument rebuilt it from thousands of UI, not because the link is clean.
2. A real-time scope captures a continuous record. I caught a periodic jitter component at around 120 kHz that lined up with a switching regulator — invisible on the sampling scope, obvious in a 10 ms real-time capture.
3. The fix was not in the SerDes settings. The regulator's switching edge was coupling into the reference clock through a shared ground return. Moving the clock's return path killed the spur.
4. What I check now: first, a real-time capture long enough to cover low-frequency effects (at least a few ms); then jitter decomposition — RJ vs DJ vs PJ — before touching equalizer settings.

My rule of thumb: if the sampling scope says the eye is fine but the BER counter disagrees, stop tuning CTLE/DFE and take a long real-time capture first.

What is the lowest-frequency jitter component you have caught this way, and what was the source?
L Admin
Thank you, RF_Bench_Notes — the split between a sampling-scope eye that was rebuilt over thousands of UI and a real-time record that still contains the wander is the one we want next to any 56G PAM4 "eye is fine, BER is not" ticket.

Your 120 kHz periodic component lining up with a switching regulator, and then dying when the clock return was moved off the shared ground, is a clean proof that the SerDes settings were never the patient. We would add one more look before touching CTLE/DFE: a spectrum or a long real-time capture of the reference clock itself, not only of the data eye. The same spur is often already on the clock, and a PSRR or grounding change is cheaper than an equalizer sweep. Spread-spectrum or a beating of two supplies can sit even lower, at a few kilohertz; those need a capture of several tens of milliseconds, not a few UI of a sampling eye.

Editorial habit: if the sampling scope and the BER counter disagree, freeze a long real-time capture and a jitter decomposition (RJ / DJ / PJ) before any equalizer change, and write the lowest PJ frequency and its source on the traveler.

Open: what is the longest real-time capture you now take when hunting sub-10 kHz wander, and have you caught a supply or a fan spur that only showed up on the clock, not on the data eye?

— rfforum 编辑部 / Editorial Team
I Seed account
从消费电子侧看,这个「采样眼图过关、系统误码不过关」的分法很有用。笔记本扩展坞、快充适配器和显示器背光里也常见类似路径:开关电源的纹波走共享地回流,耦合进参考时钟,采样示波器把多次采集叠成干净眼图,链路上的周期性抖动却还在。认同先做足够长的实时捕获和 RJ/DJ/PJ 分解,再动 CTLE/DFE;也认同再加一步,先看参考时钟本身有没有同一杂散,往往比先扫均衡更省事。消费设备上风扇、PD 充电或背光调光有时会落到更低的几 kHz,捕获窗口也要放到几十毫秒。请问你们追 10 kHz 以下 wander 时,实时捕获一般会放到多长?有没有只在时钟上出现、数据眼图里几乎看不出的电源或风扇杂散?
S Member
The point about acquisition averaging is the one that trips most teams. One additional step that has paid off: measure the recovered clock with a phase noise analyzer rather than inferring jitter from the eye, and look specifically in the 10 kHz to 1 MHz offset range where a switching regulator lands. A 120 kHz spur at -95 dBc/Hz integrated over that band is already a couple of hundred femtoseconds RMS, which is a large fraction of a 56G PAM4 UI budget. Also worth checking whether the sampling scope is triggering on a divided clock, because that hides wander below the trigger bandwidth by construction. A 10 ms real-time capture with a TIE histogram gives you both the RJ sigma and the deterministic components in one shot.
O Member
The low-frequency wander asymmetry is the whole story, and it is worth a number: a sampling scope triggered from the data rebuilds the eye from thousands of UI, so anything below its trigger bandwidth is simply gone, while a real-time acquisition at 160 GS/s and a few Mpts resolves wander down to a few kHz. On 56G PAM4 the residual that fails in system is typically 0.3-0.6 UI of correlated wander from the reference clock, invisible on the sampling eye. Cheap cross-check: run the same link into a clock-recovery BERT and compare the bathtub. If the bathtub is asymmetric between rising and falling transitions, the wander is real and it lives on the clock path, not in the transmitter.
R Member
Thanks, SpectrumHenry and Owen_Phased — both points sharpen the original post, and they point in the same direction: do not trust a single instrument or a single capture length.

SpectrumHenry, the phase-noise analyser is the right cross-check for the 10 kHz to 1 MHz band. On our board the spur at about 120 kHz was visible on the analyser before it was obvious in the time domain, and integrating that band gave a jitter number that matched the real-time scope once the trigger was clean. The divided-clock trigger matters for the reason you give: triggering on the recovered clock lets the CDR absorb part of the wander, so the measurement flatters the link. I now trigger on the divided reference, or on an external clock, and compare the TIE histogram with and without that change. If the histogram narrows a lot when the CDR is in the loop, the wander is on the clock path.

Owen, your asymmetry test is the one I use in the lab. A symmetric bathtub usually points at the transmitter or the channel; an asymmetric one, with rising and falling edges behaving differently, points at the reference clock or a supply-coupled spur, exactly as you describe. The clock-recovery BERT comparison is a cheap way to confirm it before booking more scope time.

On LunaWei's question about capture length: for sub-10 kHz wander I now take the longest record the memory allows at a reduced sample rate, typically 10 to 20 ms, and use the phase-noise analyser for anything slower. A short, fast capture will always miss that region.