Phase-stable cable: 3 bends moved phase 4.9°, 20 °C moved it 1.9° — measure them

Started by rf_zhou · 2026-11-22 08:47 UTC · 31 views · 4 replies
R Member
A 1 m phase-stable assembly, 5.05 ns delay, 18 GHz test tone, same VNA, same 0.9 N·m torque on both ends. Over one week we saw ±6° drift and started replacing cables. It was two different effects stacked, and neither was a bad cable.

Temperature first, because it is cheap and repeatable: 25 °C → 45 °C, 20 min soak at each point, phase moved +1.9°. That is 3 ppm/°C on the electrical length — normal for a PTFE dielectric assembly, and it is a property of the cable, not a defect. If your spec says 2 ppm/°C and you measure 3, you are not buying a better cable, you are buying a different dielectric.

Bending second, but with a trap: measure it two ways. Bend the assembly three times over a 40 mm radius and measure while bent → +7.4°. Straighten it back to the original route and measure again → +4.9°, i.e. 0.76 ps did not come back. The residual is what matters for a test rack, because after a re-route you never reproduce the original shape. If you only ever measure the "straightened" number you will underestimate a fixture that gets re-routed daily.

Connector third: 5 mate/unmate cycles with a torque wrench gave 0.35°, but the same 5 cycles done by hand gave 1.1°. That is the cheapest 0.7° you will ever remove.

Order that works: fix the reference plane and frequency first → measure the temperature coefficient at two points with a soak → then do bend-and-return as a separate experiment, reporting both the bent and the returned number → only then decide whether the assembly or the fixture is at fault. The reason to do it in this order is resolution: a 0.76 ps residual is below the noise of a single sweep, so each step needs its own repeat count, 10 sweeps averaged.

One thing not to do: infer phase drift from S21 magnitude. On a 1 m assembly the ripple period is ~0.2 GHz and a 6° phase move shows up as a 0.02 dB change in the passband — inside the noise.

Question: when you qualify a phase-stable assembly for a test rack, do you specify the returned-after-bend number or only the straight number? And what bend radius do you actually write into the acceptance — 40 mm, or the radius the rack actually forces?
L Admin
Thank you, rf_zhou — splitting a week of ±6° into a 1.9° temperature move and a 4.9° residual after bend-and-return, instead of condemning the cable, is the order we want on any phase-stable rack qualification.

Editorial keepers:

1. Treat 3 ppm/°C on a PTFE assembly as a dielectric property, not a defect. If the spec is 2 ppm/°C, you are buying a different dielectric, not a "better" cable of the same type.

2. Qualify with both numbers: phase while bent, and phase after return to the original route. The 0.76 ps that did not come back is what a fixture that gets re-routed daily will actually keep. A straight-only number underestimates that residual.

3. Write the acceptance radius as the tightest bend the rack actually forces, not only a convenient 40 mm lab coupon. Then fix torque: five hand mates cost you 0.7° that a wrench removes for almost nothing.

4. Do not infer phase drift from S21 magnitude. On a 1 m assembly a 6° move is about 0.02 dB of passband ripple, inside the noise. Each step needs its own repeat count (your 10-sweep average) because 0.76 ps sits below a single sweep.

Open: besides the returned-after-bend number, do you also record the radius the rack actually forces on the acceptance sheet, or only the lab coupon radius?

— rfforum 编辑部 / Editorial Team
F Member
3 ppm/degree C is what a solid PTFE dielectric gives you, and the 20-minute soak is the step most people skip, which is why their numbers scatter. Two additions. Separate elastic from plastic by measuring the same radius in both bend directions: a cable that returns to +4.9 degrees after one bend and +5.1 after five at the same radius is behaving elastically, and that 0.76 ps is a genuine electrical length change from conductor and shield geometry rather than a connector problem. Second, run the temperature ramp with the assembly in its final route. A 1 m assembly dressed in a 40 mm loop carries 2-3 cm of shield-to-dielectric slip that shows up as an extra 0.3-0.5 degrees per 10 degrees C, easily mistaken for a dielectric spec miss.
R Member
I specify the returned-after-bend number, and I write the acceptance radius as the tightest bend the rack actually forces — not the 40 mm figure from the datasheet. The reason is practical: the rack is where the cable lives. A cable that passes at 40 mm but fails at the 25 mm bend behind a panel is a cable that fails in service, and the qualification paperwork will not save it. So our traveler lists two radii: the datasheet radius for reference, and the as-installed radius measured with a simple radius gauge at the tightest point in the route.

I record both numbers, though. The phase while bent tells me how much the measurement moves when the cable is disturbed during a test, and the returned number tells me whether the disturbance left a permanent set. If the while-bent number is large but the returned number is small, I treat the cable as usable but fragile — it will not tolerate being re-routed between measurements. If the returned number itself is growing with each bend cycle, the cable is done, and no amount of re-dressing will bring it back.

Two details I would add to your procedure. First, run the temperature ramp with the cable in its final route, as Frank suggested, because shield-to-dielectric slip adds a term that a straight-cable ramp will never show. Second, log the torque-wrench value and the connector serial numbers with the phase data. Your 0.35-degree versus 3.5-degree contrast is exactly the kind of result that disappears if someone re-mates by hand the next morning, and the log is what lets you prove the measurement rather than repeat it.
L Admin
Thanks, RF_Bench_Notes — specifying the returned-after-bend number at the as-installed radius, not the datasheet radius, is a very practical rule, and recording both radii on the traveler is easy to adopt.

The interpretation you give is useful: a large while-bent number with a small returned number means usable but fragile, while a returned number that grows with each bend cycle means the cable has taken a permanent set. That growth with cycles is the shield or dielectric slipping, and it does not recover.

The two additions are worth repeating: run the temperature ramp with the cable in its final route, and log the torque value and connector serials with the phase data. rf_zhou's original contrast in this thread (1.9 degrees over 20 C versus 4.9 degrees after bends) shows how easily a re-mate the next morning can erase the evidence.

Readers: do you use a radius gauge in the rack, or rely on the cable routing drawing?

— rfforum 编辑部 / Editorial Team