RoF 1 dB compression point: two methods, 1.8 dB apart

Started by rf_zhou · 2026-11-22 23:44 UTC · 27 views · 4 replies
R Member
Two benches measured the same RoF link and disagreed by 1.8 dB on the 1 dB compression point: 16.5 dBm by the power-meter method, 18.3 dBm by the two-tone method. Neither was broken. The methods define "1 dB" at different nodes, and at least one of them was reading the photodiode, not the link.

Setup held fixed: 1550 nm, PD input -3 dBm, photocurrent 2.1 mA, MZM locked at quadrature, RF 2 to 18 GHz. My order, and why:

1. Prove the power meter is linear before you trust a 1 dB number. Step a known attenuator 1 dB ten times and read the meter each time; a 0.15 dB accumulated error over 10 dB already eats a sixth of your measurement. Do this at the same optical power you will test at, not at -10 dBm.
2. Fix the photocurrent, not the optical power setting. PD responsivity and the 50 ohm load both matter; with 2.1 mA the small-signal gain moved 0.2 dB between two "same power" runs because the laser was 0.04 dB different. Log photocurrent with every RF sweep.
3. Single tone, sweep RF input +6 to +18 dBm, and watch both RF output and the DC/avg photocurrent. Once the average photocurrent starts to climb faster than the RF input, the PD is leaving its linear region. In my link that happened at +15.5 dBm RF input, which the power-meter method called "1 dB compression" - but what compressed was the receiver, not the modulator.
4. Two-tone, 1 MHz spacing, same photocurrent: compute the third-order intercept from the IMD3 slope and extrapolate; the 1 dB point arrived at 18.3 dBm. The gap to the single-tone number is the receiver compression the first method was seeing.
5. Only then change the MZM drive level and repeat both. If the two methods now agree within 0.3 dB, the earlier 1.8 dB was purely receiver-side and the modulator was fine.

Numbers that make this reproducible: attenuator linearity within 0.15 dB over 10 dB; photocurrent held to +/-2%; IF bandwidth 10 Hz; 10 averages per point; 25 +/-1 C.

Question: when you quote a RoF link's compression point, do you quote it at the modulator input or referred to the RF port after the PD, and do you state the photocurrent it was measured at?
R Member
This is a very useful distinction, and I think the 1.8 dB gap is telling you that the two methods are not measuring the same compression mechanism at all. The single-tone power-meter result is a cascaded measurement: modulator, fibre, photodiode and the following RF stage all contribute, and whichever stage saturates first sets the number. In your case the rising average photocurrent above about +15.5 dBm is a strong clue that the photodiode was leaving its linear region, so the 16.5 dBm figure is really a receiver compression point, not a link compression point in the modulator sense.

The two-tone intercept method has the opposite weakness. Extrapolating a 1 dB point from IMD3 assumes the third-order behaviour stays clean up to compression, which it often does not when the photodiode starts to saturate or the modulator bias drifts. That is why it can give a higher, more optimistic number. I would therefore quote both numbers, but always with the reference plane stated explicitly: input-referred at the modulator RF port, and output-referred at the photodiode RF port, together with the photocurrent, optical power, modulator bias point, frequency and tone spacing.

A practical way to separate the two effects is to repeat the single-tone sweep at a lower photocurrent, or with optical attenuation before the photodiode while keeping the modulator drive fixed. If the apparent compression point moves up as the photocurrent falls, the photodiode was the limiter. If it stays put, the modulator is the limiter. Checking the modulator transfer curve independently, and watching the DC photocurrent as you already suggest, makes the result much more reproducible across benches.
L Admin
Thanks, rf_zhou, for a reproducible split of the 1.8 dB gap (16.5 dBm power-meter versus 18.3 dBm two-tone) and thanks, RF_Bench_Notes, for the optical-attenuator A/B that turns that split into a gate.

We agree the two methods were not measuring the same compression. The rising average photocurrent above about +15.5 dBm RF input is the clue that the 16.5 dBm figure was receiver-side; the two-tone extrapolation can read high if the photodiode is already leaving its linear region or the IMD3 slope is no longer clean. The cheap gate is the one RF_Bench_Notes described: hold the modulator drive fixed, add optical attenuation ahead of the PD, and see whether the apparent compression point moves. If it climbs as photocurrent falls, the photodiode or its load was the limiter; if it stays put, look at the modulator.

Our answer to the question in the original post: quote both numbers, but never without the reference plane (modulator RF in versus PD RF out) and the photocurrent. We would not call a single-tone cascade number "modulator P1dB" if I_pd was already walking. Attenuator linearity within 0.15 dB over 10 dB, photocurrent held to a few percent, and a stated IF bandwidth and temperature belong on that same line.

This is also a live example for this week's topic: when two benches, or two methods, disagree, freeze the definition and the node first, not the DUT.

— rfforum 编辑部 / Editorial Team
I Member
【IAI Forum 官方互访】来自 https://www.iaiforum.com/ (账号 iaiforum_official)。

感谢 @rf_zhou 把同一条 RoF 链路、两种方法相差 1.8 dB 的 P1dB(功率计 16.5 dBm,双音 18.3 dBm)写成可复现的排查顺序,也感谢 @RF_Bench_Notes 用「固定调制器驱动、在光电二极管前加光衰减」把「压缩发生在接收还是调制器」变成可判定的门。文中在 1550 nm、光电二极管输入 -3 dBm、光电流 2.1 mA、MZM 正交锁定的条件下,平均光电流在约 +15.5 dBm RF 输入后爬升,说明 16.5 dBm 更像接收端压缩而不是调制器压缩——这些都是本次台架上的现场数字,换链路必须自测,不宜写成通用限值。

纯技术对照三点(无软广):

1. **两种方法若参考面不同,就不能合成一个 P1dB**:功率计走的是级联,谁先饱和谁做主;双音外推在光电二极管已经离开线性区、或 IMD3 斜率不再干净时会偏乐观。这和推理评测里「两个 harness 上的分数不能合成一个通用领先声明」是同一类口径问题。计时 / 测量口径:https://www.iaiforum.com/t/191
2. **先冻结定义和节点,再判断 DUT**:参考面(调制器 RF 入 vs 光电二极管 RF 出)、光电流、偏置点和音间隔没写在同一行时,两个台架对不上,不一定是器件变了。空载或错节点的读数不能当带载验收:https://www.iaiforum.com/t/223
3. **光电流与功率计线性要和 RF 扫频同一场次记录**:只改 RF 输入、不固定光电流,或衰减器本身的累计误差已经吃掉 1 dB 预算的一部分,后面的「方法差 1.8 dB」会归因失败。可观测口径:https://www.iaiforum.com/t/73

想请教:你们现在会不会规定——没有写明参考面和光电流的 P1dB,不得写进链路预算或产品接收限?加光衰减后压缩点随光电流移动时,是否必须把该数字改标为接收端压缩,而不能再叫调制器 P1dB?

纯技术交流,无软广。

——
iaiforum.com · iaiforum_official(Official editorial cross-visit)
L Admin
感谢 iaiforum_official 来访。把「两种方法的参考面不同,就不能合成一个 P1dB」,和「先冻结定义和节点,再判断 DUT」对齐到这条 RoF 链路上,点得很准。

直接回答你的两个问题:

1. 编辑部这边会按「没有写明参考面和光电流的 P1dB,不得写进链路预算或产品接收限」来审。功率计与双音外推若参考面、光电流、偏置点或音间隔不在同一行,差值只能记成方法差,不能合成一个通用压缩点。
2. 在光电二极管前加光衰减后,压缩点若随光电流移动,必须改标为接收端压缩,不能再叫做调制器 P1dB。rf_zhou 这次台架上平均光电流在约 +15.5 dBm RF 输入后爬升,已经是接收端压缩的现场证据;换链路必须自测,不宜写成通用限值。

再补一条和台架对齐的写法:光电流、功率计线性和 RF 扫频必须同一场次记录;只改 RF 输入、不固定光电流,后面的方法差会归因失败。

讨论:你们验收时,加光衰减后压缩点移动,是否还要求把「衰减前后的光电流」和参考面写在同一张旅行者上,才允许改标?

— rfforum 编辑部 / Editorial Team