Phase noise at -130 dBc/Hz: direct spectrum vs cross-correlation, 7 dB apart

Started by rf_zhou · 2026-11-23 23:44 UTC · 16 views · 3 replies
R Member
Two ways of reading the same 100 MHz OCXO at 10 kHz offset gave -128.2 dBc/Hz (direct, spectrum analyzer, RBW 1 Hz) and -135.4 dBc/Hz (cross-correlation, two independent receive paths, 2x10 averages). The oscillator is fine; the 7 dB was my analyzer.

Order that settled it:

1. Measure the analyzer's own floor first. Both inputs terminated in 50 ohm, same RBW, same span: direct-mode floor at 10 kHz offset is -132 dBc/Hz on this unit, and the cross-correlation floor with the same settings is -168 dBc/Hz. A DUT reading 4 dB above the instrument floor already means a quarter of that number is the instrument.

2. Check the correction you are actually applying. RBW 1 Hz vs 10 Hz is 10 dB, and after normalizing both traces to 1 Hz they should coincide; mine did not, which is the hint that the direct trace was floor-limited rather than resolution-limited.

3. Cross-correlate with two separate LOs and equal path loss. Averaging N=100 buys about 10 dB over a single path (sqrt N); I used 2x10 and got 7 dB, consistent with correlated leakage in the splitter rather than with the theory being wrong.

4. Validate on a known source: a second OCXO with a published -140 dBc/Hz at 10 kHz. Direct mode reported -133 dBc/Hz (the floor), cross-correlation -138.6 dBc/Hz, 1.4 dB pessimistic because of the splitter's residual correlation.

Numbers I now write next to every phase-noise result: analyzer floor at the offset of interest, RBW, number of correlations, splitter loss and the reference source used. Without the floor, -128 dBc/Hz is not reproducible.

Question: does anyone publish the analyzer floor alongside the DUT number in bench notes? If not, how do you compare two labs that both report "-130 dBc/Hz at 10 kHz"?
L Member
A practical way to make the floor transparent is to run a self‑test at the exact offset you will quote. Set the analyzer to the same RBW (e.g., 1 Hz), terminate both inputs with a 50 Ω load, and record the displayed noise floor. Then repeat the measurement with a 0 dB attenuator in each path; the difference isolates the splitter loss and any residual correlation. For a typical bench‑top analyzer the floor at 10 kHz offset is around –132 dBc/Hz (1 Hz RBW), so a DUT reading 4 dB above that is effectively floor‑limited. When you publish the result, append a short note such as “Floor @10 kHz, 1 Hz RBW = –132 dBc/Hz; splitter loss = 3 dB; N‑corr = 20”. This format lets another lab subtract the reported floor and compare the true DUT contribution.

Another check is to inject a low‑phase‑noise reference (e.g., a –140 dBc/Hz OCXO) and verify that the measured value never exceeds the reference by more than the expected splitter loss. If it does, the floor is still dominating. Some groups also use the “two‑tone” method: drive the analyzer with two tones spaced 1 kHz apart, measure the intermodulation floor, and extrapolate to the single‑tone phase‑noise floor.

Do you already have a calibrated reference source that you could use to benchmark the floor across multiple instruments?
R Member
Yes — I now publish the analyzer floor alongside every phase-noise number, because without it the DUT number is not really comparable between labs. My bench-note format is one line per offset: DUT reading, analyzer floor at that offset with the same RBW and span, number of correlations, and the reference source used for validation. If the DUT is less than about 10 dB above the floor, I flag the result as floor-limited and quote it as an upper bound rather than a measurement.

Comparing two labs that both report around -130 dBc/Hz at 10 kHz only works if both also report the floor. A lab with a -132 dBc/Hz floor reporting -130 dBc/Hz is really saying the DUT is somewhere below that, while a lab with a -150 dBc/Hz floor reporting the same number has actually measured the device. The two results look identical and mean different things.

On the reference question: I keep a second OCXO with a published -140 dBc/Hz at 10 kHz as a transfer standard, and I run it at several offsets, not just 10 kHz, because the floor is offset-dependent. If the direct mode reads that reference at the floor and the cross-correlation mode reads it within 1-2 dB of its published value, I trust the setup for that day. I also repeat the floor check after any cable or splitter change, since a splitter with poor isolation can add correlated leakage and raise the apparent floor by several dB without any warning on the screen.
L Admin
Thanks, rf_zhou, for opening this with a fully worked case, and thanks to RF_Bench_Notes and luna_wei for the follow-ups. Taken together, this thread is close to a procedure.

The key numbers line up. A direct-mode floor of -132 dBc/Hz at 10 kHz and a DUT reading of -128.2 dBc/Hz means the DUT is only about 4 dB above the floor. Subtracting the floor on a power basis removes about 2.3 dB, so the "true" DUT contribution is closer to -130.5 dBc/Hz at best, and the cross-correlation result (-135.4 dBc/Hz) says even that is pessimistic. RF_Bench_Notes' rule, "less than about 10 dB above the floor, quote it as an upper bound", is the right reporting habit: at 10 dB margin, the floor adds only about 0.4 dB.

On cross-correlation: the theoretical improvement is about 5 x log10(N) dB, so N = 100 correlations gives about 10 dB and N = 20 about 6.5 dB. Your 7 dB on 2 x 10 averages is consistent with that, and the residual correlation through the splitter is the usual limit, as your known-source check (1.4 dB pessimistic) shows.

One small editorial caution on the two-tone idea mentioned above: a two-tone intermodulation check tells you about the analyser's linearity, which is useful, but it is not a substitute for a terminated-input or known-reference measurement of the phase-noise floor. We would keep the two as separate lines in the bench note.

So the reporting line we would propose for everyone: DUT reading, floor at the same offset and RBW, number of correlations, splitter loss, and the reference used for validation. Do other labs already report it this way?

— rfforum 编辑部 / Editorial Team