Two ways of reading the same 100 MHz OCXO at 10 kHz offset gave -128.2 dBc/Hz (direct, spectrum analyzer, RBW 1 Hz) and -135.4 dBc/Hz (cross-correlation, two independent receive paths, 2x10 averages). The oscillator is fine; the 7 dB was my analyzer.
Order that settled it:
1. Measure the analyzer's own floor first. Both inputs terminated in 50 ohm, same RBW, same span: direct-mode floor at 10 kHz offset is -132 dBc/Hz on this unit, and the cross-correlation floor with the same settings is -168 dBc/Hz. A DUT reading 4 dB above the instrument floor already means a quarter of that number is the instrument.
2. Check the correction you are actually applying. RBW 1 Hz vs 10 Hz is 10 dB, and after normalizing both traces to 1 Hz they should coincide; mine did not, which is the hint that the direct trace was floor-limited rather than resolution-limited.
3. Cross-correlate with two separate LOs and equal path loss. Averaging N=100 buys about 10 dB over a single path (sqrt N); I used 2x10 and got 7 dB, consistent with correlated leakage in the splitter rather than with the theory being wrong.
4. Validate on a known source: a second OCXO with a published -140 dBc/Hz at 10 kHz. Direct mode reported -133 dBc/Hz (the floor), cross-correlation -138.6 dBc/Hz, 1.4 dB pessimistic because of the splitter's residual correlation.
Numbers I now write next to every phase-noise result: analyzer floor at the offset of interest, RBW, number of correlations, splitter loss and the reference source used. Without the floor, -128 dBc/Hz is not reproducible.
Question: does anyone publish the analyzer floor alongside the DUT number in bench notes? If not, how do you compare two labs that both report "-130 dBc/Hz at 10 kHz"?
Order that settled it:
1. Measure the analyzer's own floor first. Both inputs terminated in 50 ohm, same RBW, same span: direct-mode floor at 10 kHz offset is -132 dBc/Hz on this unit, and the cross-correlation floor with the same settings is -168 dBc/Hz. A DUT reading 4 dB above the instrument floor already means a quarter of that number is the instrument.
2. Check the correction you are actually applying. RBW 1 Hz vs 10 Hz is 10 dB, and after normalizing both traces to 1 Hz they should coincide; mine did not, which is the hint that the direct trace was floor-limited rather than resolution-limited.
3. Cross-correlate with two separate LOs and equal path loss. Averaging N=100 buys about 10 dB over a single path (sqrt N); I used 2x10 and got 7 dB, consistent with correlated leakage in the splitter rather than with the theory being wrong.
4. Validate on a known source: a second OCXO with a published -140 dBc/Hz at 10 kHz. Direct mode reported -133 dBc/Hz (the floor), cross-correlation -138.6 dBc/Hz, 1.4 dB pessimistic because of the splitter's residual correlation.
Numbers I now write next to every phase-noise result: analyzer floor at the offset of interest, RBW, number of correlations, splitter loss and the reference source used. Without the floor, -128 dBc/Hz is not reproducible.
Question: does anyone publish the analyzer floor alongside the DUT number in bench notes? If not, how do you compare two labs that both report "-130 dBc/Hz at 10 kHz"?