Above roughly 26 GHz most benchtop spectrum analyzers stop using fundamental mixing and switch to harmonic mixing, and that is where phantom spurs start to appear. You sweep a 40 GHz converter, you find spurs at frequencies the DUT cannot possibly generate, the DUT vendor says the part is clean, and everyone starts arguing about the DUT instead of the analyzer.
The mechanism is worth stating precisely, because every diagnostic below follows from it. With harmonic mixing the first LO is multiplied by n and the received signal lands at n·fLO ± fIF. Because the analyzer detects magnitude, both sides of the LO harmonic produce the same IF, so an input tone at 2n·fLO − f_s also produces an IF response — that is the image. Multimode harmonic mixing adds further responses at n·fLO ± m·fIF for m > 1, which are the "multiple responses" that show up as extra spurs inside the band. None of these are DUT products, but all of them are real responses to real input tones, which is why they survive some tests and vanish under others.
Three checks that settle the question, in the order I would run them:
1) LO-harmonic tracking. Note where the analyzer changes harmonic number across your span (the band edges are documented, and instruments often indicate the harmonic in use). Tune the same signal in two adjacent harmonic bands and compare. A real signal keeps its absolute frequency and its level stays within the tolerance set by the different conversion loss and preselector. An image or multiple response appears only in part of the span, its indicated frequency jumps when the harmonic changes, and its level can change by many dB. Log the band plan and the harmonic number with every spur entry — a spur table without the harmonic number is not evidence.
2) External filter in front of the mixer. An image is caused by a real tone at the image frequency arriving at the input; attenuate or filter that tone and the image goes away, while the in-band signal survives. This is the decisive test, and it is the one people skip. A front-end attenuator is not decisive: it reduces the image and the real tone together, so a level change proves nothing.
3) Change the IF, or compare against a terminated input. Image position depends on fIF, so an image moves when the IF changes while a real tone does not. And if a spur survives with a 50 Ω termination on the input, it is internal — LO leakage, a multiplied LO harmonic, or an IF feedthrough — not a signal from the DUT.
Two secondary facts that explain measurements people misread. Close-in phase noise degrades by 20 log10(n) when the LO is multiplied, so an n = 4 band shows roughly 12 dB worse close-in noise than the fundamental band; comparing phase-noise plots from two bands without adjusting for n produces false conclusions. And frequency readout accuracy degrades in the same proportion, because the LO error is multiplied by n — a 1 ppm timebase error at 40 GHz with n = 4 is a much larger absolute number than the same error at 10 GHz.
Checklist before publishing a spur table from a harmonic band:
1) Record which harmonic band each spur was measured in, and whether a preselector was in path.
2) Re-tune the same spur in an adjacent band and note whether it moves.
3) Run the external filter (or waveguide high-pass) test and record the delta.
4) Terminate the input and confirm no internal spurs are being counted.
5) Check the IF path: a different IF setting must move an image, not a real tone.
6) Use a verified source or comb generator at a known frequency, not the DUT, as the reference for frequency accuracy.
7) Note the LO harmonic multiplier with any phase-noise or frequency-accuracy claim.
8) Keep the harmonic, IF, span and RBW in the same row as the spur level — otherwise the table cannot be reproduced or disputed.
The general rule: below the harmonic-mixing threshold a spur is probably the DUT's; above it, the analyzer is a suspect until the three checks have been run. Proving a signal is real is cheaper than arguing with a vendor about a spur that only exists inside the instrument.
The mechanism is worth stating precisely, because every diagnostic below follows from it. With harmonic mixing the first LO is multiplied by n and the received signal lands at n·fLO ± fIF. Because the analyzer detects magnitude, both sides of the LO harmonic produce the same IF, so an input tone at 2n·fLO − f_s also produces an IF response — that is the image. Multimode harmonic mixing adds further responses at n·fLO ± m·fIF for m > 1, which are the "multiple responses" that show up as extra spurs inside the band. None of these are DUT products, but all of them are real responses to real input tones, which is why they survive some tests and vanish under others.
Three checks that settle the question, in the order I would run them:
1) LO-harmonic tracking. Note where the analyzer changes harmonic number across your span (the band edges are documented, and instruments often indicate the harmonic in use). Tune the same signal in two adjacent harmonic bands and compare. A real signal keeps its absolute frequency and its level stays within the tolerance set by the different conversion loss and preselector. An image or multiple response appears only in part of the span, its indicated frequency jumps when the harmonic changes, and its level can change by many dB. Log the band plan and the harmonic number with every spur entry — a spur table without the harmonic number is not evidence.
2) External filter in front of the mixer. An image is caused by a real tone at the image frequency arriving at the input; attenuate or filter that tone and the image goes away, while the in-band signal survives. This is the decisive test, and it is the one people skip. A front-end attenuator is not decisive: it reduces the image and the real tone together, so a level change proves nothing.
3) Change the IF, or compare against a terminated input. Image position depends on fIF, so an image moves when the IF changes while a real tone does not. And if a spur survives with a 50 Ω termination on the input, it is internal — LO leakage, a multiplied LO harmonic, or an IF feedthrough — not a signal from the DUT.
Two secondary facts that explain measurements people misread. Close-in phase noise degrades by 20 log10(n) when the LO is multiplied, so an n = 4 band shows roughly 12 dB worse close-in noise than the fundamental band; comparing phase-noise plots from two bands without adjusting for n produces false conclusions. And frequency readout accuracy degrades in the same proportion, because the LO error is multiplied by n — a 1 ppm timebase error at 40 GHz with n = 4 is a much larger absolute number than the same error at 10 GHz.
Checklist before publishing a spur table from a harmonic band:
1) Record which harmonic band each spur was measured in, and whether a preselector was in path.
2) Re-tune the same spur in an adjacent band and note whether it moves.
3) Run the external filter (or waveguide high-pass) test and record the delta.
4) Terminate the input and confirm no internal spurs are being counted.
5) Check the IF path: a different IF setting must move an image, not a real tone.
6) Use a verified source or comb generator at a known frequency, not the DUT, as the reference for frequency accuracy.
7) Note the LO harmonic multiplier with any phase-noise or frequency-accuracy claim.
8) Keep the harmonic, IF, span and RBW in the same row as the spur level — otherwise the table cannot be reproduced or disputed.
The general rule: below the harmonic-mixing threshold a spur is probably the DUT's; above it, the analyzer is a suspect until the three checks have been run. Proving a signal is real is cheaper than arguing with a vendor about a spur that only exists inside the instrument.