新闻解读:electronica 2026 上 Keysight 把 PNA-X Pro 动态范围说到 170 dB:手册在 10 Hz 中频、无平均下,高增益典型值最高约 151 dB / Keysight PNA-X Pro "170 dB" at electronica: the 10 Hz IF data-sheet typ. tops out near 151 dB

Started by LunaWei · 2026-11-26 02:04 UTC · 10 views · 1 replies
L Admin
[Industry news interpretation, rfforum Editorial Team]

Sources (retrieved 26 November 2026):
- everything RF, dated 19 November 2026: "Keysight Showcases Vector Network Analyzer Portfolio and Other RF Test Solutions at electronica 2026" — https://www.everythingrf.com/news/details/23065-keysight-showcases-vector-network-analyzer-portfolio-and-other-rf-test-solutions-at-electronica-2026
- Keysight, PNA-X Pro Vector Network Analyzer data sheet (NA5202A / NA5204A / NA5205A) — https://www.keysight.com/content/dam/keysight/en/doc/ungate/data-sheets/PNA-X-Vector-Network-Analyzer.pdf

What the trade report said. At electronica 2026 in Munich, everything RF reported a walk-through of Keysight's VNA portfolio by Axel Hahn (Application Engineer, Component Test). The flagship platform discussed was the PNA-X Pro, described as combining network analysis with multiple integrated sources and receivers so that gain, phase, group delay, noise figure, compression, intermodulation and modulated-signal measurements can share one setup. In that demonstration, the instrument's dynamic range was quoted as "up to 170 dB". The same report also mentioned the E5080B ENA (the everything RF product blurb lists 2/4-port coverage from 9 kHz to 4.5/6.5/9/14/20 GHz) and noted that Keysight's latest 250 GHz VNA was not on the stand.

What the data sheet actually publishes. The official PNA-X Pro data sheet defines system dynamic range at the test port under explicit conditions: 10 Hz IF bandwidth, no averaging, and the figure does not include crosstalk. Under those conditions, the highest typical number we can find for the NA5202A (10 MHz–26.5 GHz) in high-gain mode is 151 dB, and only in the 2.234–7 GHz band (spec 146 dB in the same cell). The same table then falls with frequency: 146 dB typical from 12.55–21.25 GHz (spec 140 dB) and 140 dB typical from 24.5–26.5 GHz (spec 132 dB). Low-gain mode is lower still (135 dB typical in the 2.234–7 GHz band). We did not find a 170 dB line in that data sheet under the stated conditions. A Keysight marketing page for the same family also quotes a 140 dB dynamic-range figure, and 137 dB at 20 GHz for the NA5202A.

How two "dynamic range" numbers become incomparable. Dynamic range on a VNA is source power minus the noise floor in the receiver, in a stated IF bandwidth, in a stated receiver-gain mode, at a stated frequency, usually without averaging. On a power-density basis, narrowing the IF bandwidth from 300 Hz to 10 Hz is 10 log10(300/10) ≈ 14.8 dB of apparent floor improvement; turning averaging on adds more. Two figures taken at different IF bandwidths, or one taken in a demonstration trace and one taken from the warranted table, are not the same test. The 170 dB phrase is therefore not a number we would copy onto a traveler without the IF bandwidth, gain mode, frequency and averaging state written on the same line.

What we would write before quoting a dynamic-range or isolation figure from a VNA:
1. Instrument model and receiver-gain mode (low / high / auto)
2. IF bandwidth, averaging, and whether crosstalk is included
3. Frequency, and whether the number is spec, typical or a single-unit demonstration
4. How the cal kit and the test-port cables were included, because a 150 dB system figure is not the isolation you will see through a pair of 24 GHz cables and a poorly torqued adapter

Discussion: when you sign an isolation or a filter-rejection number "measured on a PNA-class VNA", which IF bandwidth and gain mode do you write next to it — and have you ever had a 10–20 dB argument that disappeared once both labs used the same IF bandwidth?

— rfforum 编辑部 / Editorial Team
L Admin
Editorial addendum — a number to put on the traveler with any "VNA dynamic range" or isolation claim.

The 10 Hz IF bandwidth used in the PNA-X Pro system-dynamic-range table is about 14.8 dB quieter, on a power-density basis, than a 300 Hz IF (10 log10(300/10)). If lab A signs a 120 dB isolation at 10 Hz IF and lab B repeats the same filter at 1 kHz IF with a different receiver-gain mode, a 10–20 dB disagreement can be the instrument setup rather than the DUT. We would like to hear from readers who have already had to reconcile two isolation numbers after aligning IF bandwidth and gain mode — what residual was then left, and was it the cables or the cal kit?

— rfforum 编辑部 / Editorial Team