Y-factor noise-figure uncertainty: ENR tolerance, the Y ratio, and the 0.2 dB that belongs to the standard

Started by LunaWei · 2026-11-26 02:05 UTC · 20 views · 4 replies
L Admin
[FAQ explainer / Learning, rfforum Editorial Team]

A noise-figure meter or a Y-factor setup that repeats to 0.05 dB is not automatically accurate to 0.05 dB. A large part of the uncertainty was already on the excess-noise-ratio (ENR) label of the noise source, and it transfers into the result almost one-for-one when the Y-factor is large.

1. The Y-factor in one line
The DUT is measured once with the noise source on (hot) and once with it off (cold, usually a 50 Ω termination at T0 = 290 K, or the source's own cold state). Y = N_hot / N_cold. For a source with excess noise ratio ENR (linear), the DUT's noise factor is F = ENR / (Y − 1), after you have also accounted for the second-stage contribution if you are not measuring a high-gain block. In decibels, NF = 10 log10 F.

2. How a 0.2 dB ENR tolerance becomes a 0.2 dB NF error (worked numbers)
Suppose ENR = 15.0 dB (linear 31.62) and the measured Y is 12.0 dB (linear 15.85). Then F = 31.62 / (15.85 − 1) = 2.13, so NF ≈ 3.27 dB. If the ENR calibration is actually 15.2 dB (linear 33.11) and you still use 15.0 dB, the true F is 33.11 / 14.85 ≈ 2.23, NF ≈ 3.48 dB. You have a 0.21 dB error in NF from a 0.2 dB ENR error, and no amount of averaging the Y-factor will remove it. When Y is large (a high-gain, low-NF DUT), ∂NF/∂ENR ≈ 1. When Y is small (a poor-NF or low-gain DUT), the transfer is smaller, but then the Y-factor itself is noisier and the second-stage correction grows.

3. The other terms that belong in the same uncertainty budget
- Second-stage (or receiver) noise factor, and the available gain used to refer it. A 0.15 dB error in the measured DUT gain becomes a first-order error in the Friis correction when the DUT gain is only 10–15 dB.
- Mismatch: the noise source, the DUT and the receiver do not all present 50 Ω. A source with |Γ| = 0.15 and a DUT input with |Γ| = 0.2 can move the delivered ENR by a few tenths of a dB unless a tuner or a well-specified pad is used and included in the correction.
- Temperature of the cold state. If "cold" is not 290 K, the ENR definition and T_cold have to be consistent; a 5 K error on T_cold is already visible at the 0.05–0.1 dB level on a low-NF LNA.
- Bandwidth and linearity of the receiver. A Y-factor taken with the receiver in compression, or with an LO spur inside the noise bandwidth, is not a Y-factor.

4. What we would write next to a signed noise-figure figure
- Method (Y-factor, cold-source, or a noise-figure analyser), IF / RBW, and averaging
- Noise-source model, the ENR table frequency actually used, the calibration date, and the stated ENR uncertainty
- Second-stage correction, DUT gain, and whether a pad was included
- T_cold and the mismatch correction, or a statement that none was applied

Discussion: on your last LNA sign-off, was the 0.2 dB you argued about the Y-factor repeatability, the ENR label, or a mismatch / second-stage term that was never written down?

— rfforum 编辑部 / Editorial Team
L Admin
Editorial addendum — a question for anyone who has argued 0.1–0.2 dB on an LNA noise figure.

Was the disagreement still there after both sides used the same ENR table, the same T_cold and the same second-stage correction? If it vanished, the 0.2 dB belonged to the standard, not to the LNA. If it remained, what term in the mismatch or gain correction finally closed it?

— rfforum 编辑部 / Editorial Team
R Member
One line worth adding to the table: how much of the 0.05 dB repeat actually survives into NF.

From F = ENR/(Y-1), the Y-side sensitivity is ΔNF(dB) ≈ 0.4343 × (Y/(Y-1)) × ΔY(dB). At Y = 12 dB that factor is 15.85/14.85 = 1.067, so 0.05 dB of repeat becomes 0.053 dB. At Y = 3 dB it is 2.0/1.0 = 2.0, so 0.05 dB becomes 0.10 dB; at Y = 20 dB it is 100/99 = 1.01. In other words the Y-factor noise is essentially never the limiter once gain is decent — the ENR label is, at 1:1.

Budget we use for a 1.1 dB LNA at Y = 15 dB, all terms at k = 2:

- ENR tolerance ±0.2 dB: 0.20 dB (1:1 transfer, unavoidable without a better source)
- cold load 290 ± 2 K: 0.03 dB
- mismatch (both ports better than -14 dB return loss): 0.02 dB
- second stage, G = 28 dB / F = 8 dB: 0.06 dB
- repeatability 0.05 dB × 1.03: 0.05 dB

RSS ≈ 0.22 dB. Two labs reporting 0.9 and 1.1 dB on the same part are therefore inside one budget, and the disagreement does not belong to the LNA unless the ENR tables and the cold termination differ.

A practical check we do before arguing: cross-calibrate the two noise sources against each other on the same DUT with the same analyser, and log the ENR table revision in the report. If the 0.2 dB survives that, it is the standard.

Question: does anyone correct the ENR table for the actual hot-source temperature when the DUT ambient is not 23 °C? At 15 dB ENR the correction is small, but at 6 dB sources we have seen 0.1 dB from thermal drift alone.
R Member
rf_zhou, yes — I do correct the ENR table for the actual hot-source temperature, and the 0.1 dB drift you saw at 6 dB ENR is about the size I would expect, which is why I no longer treat it as negligible on a low-ENR source.

The correction is straightforward once the temperatures are written down. The ENR table is quoted for a cold state of 290 K, so the effective ENR at the actual cold temperature T_cold is ENR_eff = ENR_table + 10 log10(290 / T_cold) in the hot-minus-cold sense, and the hot temperature enters through the same ratio: a source whose body has warmed from 23 C to 35 C has its hot noise temperature up by about 12 K, which on a 6 dB ENR source is roughly 0.08 to 0.12 dB of ENR shift, exactly the order you measured. On a 15 dB source the same 12 K is a smaller fraction of the excess, so the shift drops to a few hundredths of a dB and is usually buried in the ENR tolerance — which is why the correction matters most on the low-ENR sources that are otherwise attractive for low-NF DUTs.

In practice I measure the source body temperature with a small thermocouple taped to the source case, log it with the Y-factor, and let the source soak for at least 30 minutes after switch-on before taking the data. I also record the ambient temperature at the DUT, because a DUT at 35 C is not the same cold state as a DUT at 23 C, and the T_cold used in the calculation has to match the DUT's actual temperature, not the room's nominal value. If the two temperatures differ by more than about 5 K, I correct both sides and state both temperatures next to the result. The correction does not remove the ENR tolerance — that 0.2 dB is still there — but it removes a systematic drift that would otherwise masquerade as a DUT difference between a morning and an afternoon measurement.
L Admin
rf_zhou, RF_Bench_Notes — the two follow-ups together are the uncertainty line we would put on a Y-factor report.

rf_zhou's budget for a 1.1 dB LNA at Y = 15 dB (k = 2), RSS ≈ 0.22 dB and dominated by the ±0.2 dB ENR tolerance, with the 0.05 dB repeat transferring almost 1:1 at that Y, is why two labs reporting 0.9 dB and 1.1 dB on the same part are still inside one standard. The disagreement does not belong to the LNA unless the ENR tables or the cold termination differ. Cross-calibrating the two sources on the same DUT and logging the ENR-table revision is the check we would also run before arguing.

RF_Bench_Notes — yes, we would correct for the actual source-body and DUT-side cold temperatures, and we would soak the source (your 30 minutes after switch-on) before taking the Y-factor. The 0.1 dB order you and rf_zhou saw on a 6 dB ENR source, versus only a few hundredths on a 15 dB source for the same 12 K walk, is why the correction matters most on the low-ENR heads that are otherwise attractive for low-NF DUTs. We would use the manufacturer's ENR-versus-temperature note, or a measured cross-cal, for the size of that correction, rather than assuming that a 12 K change in case temperature is a 12 K change in excess noise.

Both belong on the same line: ENR table revision, source-body temperature, DUT T_cold, and whether a cross-cal was done. The 0.2 dB ENR tolerance is still there; the temperature line only removes a systematic that would otherwise look like a morning-versus-afternoon DUT difference.

Open question: do both labs already write the source-body temperature on the report, or only the ENR-table revision?