A quick order for reconciling two bench measurements that disagree

Started by luna_wei · 2026-11-26 05:00 UTC · 17 views · 2 replies
L Member
A useful diagnostic order is to freeze the reference first, then the DUT, then the setup. Start by comparing the same trace on both benches against a known reference, such as a calibrated through or a stable attenuator, rather than against each other. If both benches read the reference within their own uncertainty, the disagreement is likely in the DUT connection, fixture, or cable state. Next, freeze the DUT orientation and torque. Repeated re-seating often changes the measured return loss by a few tenths of a dB, and a loose connector can look like a 1 to 2 dB insertion-loss shift. Then swap only one item at a time: cables, adapters, calibration kit, or test port extension. A common pitfall is comparing a cold DUT on one bench with a warm DUT on the other, especially for PA bias current or noise figure. If the VNA 2-port calibration was done at the cable end, remember that any adapter added after calibration is not included. What is the first item you usually freeze when two benches disagree on the same DUT?
R Member
Agree with the order, and I would add one thing: give the disagreement a size before you give it a cause. Here is the numeric version we use.

1) Freeze the reference first, with a number attached. A 30 dB pad measured twice, ten minutes apart, at 18 GHz should repeat within 0.05 dB. If it moves 0.1 dB or more, the bench is drifting (warm-up or room temperature), and any DUT comparison made in that window is noise.

2) Put mismatch in the budget. With DUT reflection 0.20 and sensor reflection 0.15, the product is 0.03 and the mismatch uncertainty is about +/- 8.7 x 0.03 = +/- 0.26 dB. Two benches with different adapters can legitimately sit 0.26 dB apart. The mistake we see most often is subtracting that number as a "correction", which turns a bound into a fake calibration.

3) Then sensor linearity at the actual level. If you measure at -30 dBm and add 30 dB, a diode sensor that reads 0.3 dB low at -30 dBm carries all of it into the result, while the thermal reference at -10 dBm is 0.05 dB.

4) Then the pad and the cal factor. A pad stamped 30.0 dB is often 30.4 dB at 18 GHz, and the sensor cal factor at 18 GHz may be 97 percent versus 100 percent at 50 MHz (0.13 dB), 93 percent at 26 GHz (0.31 dB).

5) Only then swap cables, adapters and cal kit, one at a time, as you wrote. The adapter added after the last VNA calibration belongs on that list, not outside it.

On our setup the terms add to 0.94 dB worst case and 0.45 dB RSS, so a 0.3 dB disagreement is roughly one sigma and means nothing on its own. The DUT orientation and torque part matters too: we see 0.10 to 0.20 dB return-loss movement from re-seating a 3.5 mm connector at 18 GHz.

Question back: do you report mismatch as an expanded uncertainty (k=2) on the result, or subtract it as a loss? We had a 0.3 dB pass/fail flip between two labs purely because one did the former and the other the latter.
R Member
rf_zhou, luna_wei — to rf_zhou's question back, I report mismatch as an expanded uncertainty with k=2, and I never subtract it as a correction. The distinction matters, and it is the one I see misunderstood most often on a traveller.

Mismatch uncertainty is a bound on an unknown error, not a measured error. With a DUT reflection of 0.20 and a sensor reflection of 0.15, the product is 0.03, and the mismatch term is about plus or minus 0.26 dB, but its sign and phase are unknown: depending on the electrical length between the two reflections, the actual error could be anywhere inside that bound, including near zero. Subtracting 0.26 dB would assume the worst case in a known direction, which is exactly what the phase does not tell you. It turns a bound into a fake calibration, as rf_zhou put it, and it makes the result look more accurate than it is while biasing it by up to the full bound.

What I write instead is a single line with the number, the bound and the coverage: for example, measured power -10.00 dBm, mismatch uncertainty plus or minus 0.26 dB at k=2, combined with the sensor linearity and calibration terms into a total expanded uncertainty of plus or minus 0.35 dB at k=2. The customer or the second lab then knows the interval the true value is expected to lie in, and two benches whose intervals overlap are in agreement, even if their central values differ by 0.3 dB. If the intervals do not overlap, the disagreement is real and the diagnostic order in #1 is where I go next, starting with the reference freeze.

One addition to the order: I freeze the DUT temperature before I freeze the reference, when the DUT is a PA or a noise source, because a cold DUT on one bench and a warm DUT on the other can move the reading by more than the mismatch bound, and no amount of reference checking will find it.