Two power readings that disagree by 0.3 dB at 18 GHz: mismatch uncertainty first, then sensor linearity

Started by rf_zhou · 2026-11-26 06:02 UTC · 15 views · 3 replies
R Member
Same CW source, 18 GHz, nominal -3.2 dBm. Bench A reads -3.19 dBm, bench B reads -3.48 dBm. Both sensors were calibrated three months ago and both benches pass their daily reference check, so the argument starts about which bench is "wrong". Before anyone touches a sensor, put a size on each term. In our case the 0.3 dB was inside the combined uncertainty, and neither bench was broken.

Step 1 - mismatch is an uncertainty, not a loss.
Bench A: source/DUT reflection 0.20 (VSWR 1.5), sensor reflection 0.15. The product is 0.03, and the mismatch uncertainty is about +/- 8.686 x 0.03 = +/- 0.26 dB. That is a bound on what the true power could be, not something you subtract. Two benches with different adapters can sit 0.26 dB apart and both be right.

Step 2 - linearity at the level you actually measure.
Our -3.2 dBm reading came from a -33 dBm point plus a 30 dB pad. A diode sensor compared against a thermal reference read 0.31 dB low at -30 dBm and 0.05 dB low at -10 dBm. Measuring at -30 dBm and adding 30 dB carries that 0.31 dB straight into the answer, plus the pad's own error.

Step 3 - calibration factor per frequency.
The datasheet cal factor at 18 GHz is 97 percent versus 100 percent at 50 MHz, which is 0.13 dB if you forget to apply it. At 26 GHz the same sensor is at 93 percent, 0.31 dB. A single "sensor is calibrated" line in the traveller is not enough; the factor has to be entered at the test frequency.

Step 4 - the plumbing between the two benches.
Two N-to-3.5 mm adapters at 0.08 dB each, a 1.2 m cable at 0.9 dB insertion at 18 GHz, and the adapter that was added after the last VNA calibration. This is the item that most often differs between two benches that otherwise look identical.

Step 5 - temperature and warm-up.
Diode drift around 0.015 dB per degree C, a 5 degree C room difference is 0.075 dB. A sensor that has been powered for 10 minutes is not the same as one powered for 60; the zero drift after the first half hour is typically 0.02 to 0.05 dB.

Budget: 0.26 (mismatch, k=1) + 0.31 (linearity) + 0.13 (cal factor) + 0.16 (adapters) + 0.08 (temperature) = 0.94 dB worst case, 0.45 dB RSS. A 0.3 dB disagreement therefore tells you nothing until the terms are separated; it is roughly one sigma of a normal setup.

What I now write on the traveller: reflection of DUT and sensor measured at the test frequency, cal factor entered, pad attenuation measured at the test frequency (30.4 dB at 18 GHz, not the 30.0 dB stamped on it), adapters counted with their own insertion loss, and the sensor warm-up time. Only then does a difference get investigated as a fault.

Question: do you report mismatch as an expanded uncertainty (k=2) on the result, or do you subtract it as a correction? We have seen a 0.3 dB pass/fail move purely because one lab subtracted and the other did not.
L Admin
rf_zhou — this is the right order, and the 0.3 dB disagreement sitting inside a 0.45 dB RSS budget is exactly why the first question should be how each term was entered, not which bench is broken.

On your question: when you only have |Γ| at the test frequency, report mismatch as an uncertainty contribution, not as a correction you subtract from the reading. The ±0.26 dB you wrote (8.686 × |Γs||ΓL|) is a bound on the unknown relative phase. With unknown phase it is usually treated as a U-shaped Type B term; the standard uncertainty is then that half-width divided by √2, about 0.18 dB in your example. The k=2 expansion belongs to the combined standard uncertainty of the whole budget, not to the mismatch term alone, and not as a one-sided 0.26 dB or 0.52 dB deduction on the power reading.

A signed mismatch correction is appropriate only when both complex reflection coefficients are measured at the test frequency, at the same reference plane as the power measurement. After that correction, what remains on the traveler is the residual uncertainty of those Γ measurements and of the connectors, which should be much smaller than 0.26 dB. If either lab subtracted the scalar bound instead, that alone can move a 0.3 dB pass/fail even when both sensors are fine.

We would write one of these two lines, never a mix of the two:

1. Mismatch: uncertainty only. |Γs| and |ΓL| at 18 GHz, the half-width, the standard uncertainty, and a note that this term was not applied as a correction.

2. Mismatch: corrected from measured complex Γ at 18 GHz, with the residual uncertainty after correction.

The rest of your traveler (linearity at the actual -30 dBm level, cal factor at 18 GHz, pad attenuation measured rather than stamped, adapters counted, sensor warm-up) is what then decides whether the remaining 0.3 dB is still a fault.

Other readers: when two labs already agree not to subtract the scalar bound, has the 0.3 dB closed after aligning the pad and the cal factor at the test frequency, or does the adapter stack still dominate?

— rfforum 编辑部 / Editorial Team
R Member
rf_zhou, LunaWei — to the open question at the end of #2, yes, the 0.3 dB has closed a disagreement for me, but only after we stopped arguing about which bench was right and ran the swap test that separates a bench error from a DUT error.

The case was two benches disagreeing by about 0.35 dB at 12 GHz on a power amplifier output. Both benches had passed their daily checks, both sensors were in calibration, and the mismatch bounds alone covered the difference, so neither side would concede. What closed it was swapping the sensors, not the benches: each sensor was moved to the other bench, with the same cables and adapters, and the readings followed the sensors, not the benches. That pointed at a linearity error at the actual measurement level on one sensor, which a later check against a thermal reference confirmed at about 0.25 dB low at that level. Once the sensor was corrected, the two benches agreed within 0.08 dB, inside the mismatch uncertainty, and the argument ended.

Two practical points from that experience. First, a swap test is cheap and decisive in a way that a budget table is not: if the difference follows the sensor, it is the sensor; if it follows the bench, it is the bench plumbing or the calibration; if it follows neither, it is the DUT or the source drifting between measurements. Second, I now write the traveller line as LunaWei suggests — mismatch as an uncertainty with the measured magnitudes, the half-width and the coverage factor — and I add one more line, the swap-test result, because a customer or a second lab trusts a measured swap far more than a computed bound. The bound tells you the disagreement is plausible; the swap tells you where it lives.
L Admin
RF_Bench_Notes — the sensor swap that closed a 0.35 dB disagreement at 12 GHz is the decisive test, and it is cheaper than another round of arguing about which budget is tighter. The difference followed the sensors, not the benches; a later check against a thermal reference found one sensor about 0.25 dB low at that level; after the correction the two benches agreed within 0.08 dB, inside the mismatch uncertainty. That is the sentence a second lab will actually trust.

One addition we would write next to the swap: do it at the actual measurement level, not only at the daily-check power. A sensor can pass a mid-range check and still be 0.2–0.3 dB off at a PA output level, which is exactly where these arguments start. The traveler then has two lines, as you suggest: mismatch as an uncertainty (magnitudes, half-width, coverage factor), and the swap-test result (followed the sensor / the bench / neither).

Open question to other readers: when you swap, do you also move the adapter and cable stack with the sensor, or only the head? A 12 GHz disagreement that follows a "sensor" is sometimes still the adapter that travelled with it.

— rfforum 编辑部 / Editorial Team