Same CW source, 18 GHz, nominal -3.2 dBm. Bench A reads -3.19 dBm, bench B reads -3.48 dBm. Both sensors were calibrated three months ago and both benches pass their daily reference check, so the argument starts about which bench is "wrong". Before anyone touches a sensor, put a size on each term. In our case the 0.3 dB was inside the combined uncertainty, and neither bench was broken.
Step 1 - mismatch is an uncertainty, not a loss.
Bench A: source/DUT reflection 0.20 (VSWR 1.5), sensor reflection 0.15. The product is 0.03, and the mismatch uncertainty is about +/- 8.686 x 0.03 = +/- 0.26 dB. That is a bound on what the true power could be, not something you subtract. Two benches with different adapters can sit 0.26 dB apart and both be right.
Step 2 - linearity at the level you actually measure.
Our -3.2 dBm reading came from a -33 dBm point plus a 30 dB pad. A diode sensor compared against a thermal reference read 0.31 dB low at -30 dBm and 0.05 dB low at -10 dBm. Measuring at -30 dBm and adding 30 dB carries that 0.31 dB straight into the answer, plus the pad's own error.
Step 3 - calibration factor per frequency.
The datasheet cal factor at 18 GHz is 97 percent versus 100 percent at 50 MHz, which is 0.13 dB if you forget to apply it. At 26 GHz the same sensor is at 93 percent, 0.31 dB. A single "sensor is calibrated" line in the traveller is not enough; the factor has to be entered at the test frequency.
Step 4 - the plumbing between the two benches.
Two N-to-3.5 mm adapters at 0.08 dB each, a 1.2 m cable at 0.9 dB insertion at 18 GHz, and the adapter that was added after the last VNA calibration. This is the item that most often differs between two benches that otherwise look identical.
Step 5 - temperature and warm-up.
Diode drift around 0.015 dB per degree C, a 5 degree C room difference is 0.075 dB. A sensor that has been powered for 10 minutes is not the same as one powered for 60; the zero drift after the first half hour is typically 0.02 to 0.05 dB.
Budget: 0.26 (mismatch, k=1) + 0.31 (linearity) + 0.13 (cal factor) + 0.16 (adapters) + 0.08 (temperature) = 0.94 dB worst case, 0.45 dB RSS. A 0.3 dB disagreement therefore tells you nothing until the terms are separated; it is roughly one sigma of a normal setup.
What I now write on the traveller: reflection of DUT and sensor measured at the test frequency, cal factor entered, pad attenuation measured at the test frequency (30.4 dB at 18 GHz, not the 30.0 dB stamped on it), adapters counted with their own insertion loss, and the sensor warm-up time. Only then does a difference get investigated as a fault.
Question: do you report mismatch as an expanded uncertainty (k=2) on the result, or do you subtract it as a correction? We have seen a 0.3 dB pass/fail move purely because one lab subtracted and the other did not.
Step 1 - mismatch is an uncertainty, not a loss.
Bench A: source/DUT reflection 0.20 (VSWR 1.5), sensor reflection 0.15. The product is 0.03, and the mismatch uncertainty is about +/- 8.686 x 0.03 = +/- 0.26 dB. That is a bound on what the true power could be, not something you subtract. Two benches with different adapters can sit 0.26 dB apart and both be right.
Step 2 - linearity at the level you actually measure.
Our -3.2 dBm reading came from a -33 dBm point plus a 30 dB pad. A diode sensor compared against a thermal reference read 0.31 dB low at -30 dBm and 0.05 dB low at -10 dBm. Measuring at -30 dBm and adding 30 dB carries that 0.31 dB straight into the answer, plus the pad's own error.
Step 3 - calibration factor per frequency.
The datasheet cal factor at 18 GHz is 97 percent versus 100 percent at 50 MHz, which is 0.13 dB if you forget to apply it. At 26 GHz the same sensor is at 93 percent, 0.31 dB. A single "sensor is calibrated" line in the traveller is not enough; the factor has to be entered at the test frequency.
Step 4 - the plumbing between the two benches.
Two N-to-3.5 mm adapters at 0.08 dB each, a 1.2 m cable at 0.9 dB insertion at 18 GHz, and the adapter that was added after the last VNA calibration. This is the item that most often differs between two benches that otherwise look identical.
Step 5 - temperature and warm-up.
Diode drift around 0.015 dB per degree C, a 5 degree C room difference is 0.075 dB. A sensor that has been powered for 10 minutes is not the same as one powered for 60; the zero drift after the first half hour is typically 0.02 to 0.05 dB.
Budget: 0.26 (mismatch, k=1) + 0.31 (linearity) + 0.13 (cal factor) + 0.16 (adapters) + 0.08 (temperature) = 0.94 dB worst case, 0.45 dB RSS. A 0.3 dB disagreement therefore tells you nothing until the terms are separated; it is roughly one sigma of a normal setup.
What I now write on the traveller: reflection of DUT and sensor measured at the test frequency, cal factor entered, pad attenuation measured at the test frequency (30.4 dB at 18 GHz, not the 30.0 dB stamped on it), adapters counted with their own insertion loss, and the sensor warm-up time. Only then does a difference get investigated as a fault.
Question: do you report mismatch as an expanded uncertainty (k=2) on the result, or do you subtract it as a correction? We have seen a 0.3 dB pass/fail move purely because one lab subtracted and the other did not.