L
Admin
[Technical note, rfforum Editorial Team]
The far-field rule 2D²/λ is used so often at FR2 that it is sometimes copied onto a range sketch without the phase error it already accepts, or without checking whether D is the largest dimension that is actually radiating. At 28 GHz the wavelength is c/f = 10.707 mm. The range length then scales with the square of the antenna size:
- 10 cm AUT: 2D²/λ = 1.87 m
- 15 cm AUT: 4.20 m
- 20 cm AUT: 7.47 m
- 30 cm AUT: 16.81 m
For comparison, the same 20 cm aperture at 10 GHz (λ = 29.98 mm) only needs 2.67 m. A lab that was comfortable at X-band can run out of room as soon as the same horn moves to 28 GHz, and a compact-range or a near-field scanner is then a geometry decision, not a luxury.
What 2D²/λ actually buys. The criterion corresponds to a maximum path-length difference of about λ/16 from the centre of a uniformly illuminated aperture to its edge, which is a 22.5° phase taper over the aperture. It is a commonly accepted starting point, not a zero-error condition. If a pattern or a gain figure is being signed to a tighter phase budget — some labs work to λ/32, which doubles the range — 2D²/λ is already too short. We would write the rule that was used (2D²/λ, 2 × 2D²/λ, or a stated phase error) on the same line as the gain or the sidelobe number, because a 0.3–0.5 dB disagreement between two ranges is often this choice, not the AUT.
What "D" should be. D is the largest radiating dimension, not the flange and not the absorbing box. A 15 cm horn with a 25 cm fixture plate still has D ≈ 15 cm if the plate is not radiating; a 15 cm patch array on a 30 cm ground plane may have to be treated as the larger of the array and the currents that actually run on the ground. If you are unsure, use the larger number and say so. Using the patch width alone, then being surprised that the measured phase across a probe scan is worse than 22.5°, is a definition problem.
Two checks that are cheaper than arguing about the formula:
1. Amplitude taper and quiet-zone size. Even a range that meets 2D²/λ can still put a 1 dB amplitude taper across the AUT, which will move a gain or a beamwidth number by a few tenths of a dB. Measure the quiet zone with a probe or a small horn at the AUT position, and write the taper and the ripple next to the far-field claim.
2. A known-gain standard. A substitution horn with a calibration that covers 28 GHz, measured in the same range length, is the acceptance we would sign before a DUT gain is written down. If the standard already disagrees with its calibration by more than the stated range uncertainty, lengthening the range or moving to a compact range is the next step, not a software correction on the DUT.
A compact range or a near-field scan is the practical answer once 2D²/λ no longer fits the room. It does not remove the need to write down the effective far-field criterion, the quiet-zone ripple, and the standard-gain check. Those three lines are what make a 28 GHz gain figure comparable between two labs.
Open question: at 28 GHz, do you still use 2D²/λ as written, do you double it, or have you already moved that size of AUT to a compact range or a near-field scanner — and how large a gain disagreement have you seen between a "just 2D²/λ" range and a longer one?
— rfforum 编辑部 / Editorial Team
The far-field rule 2D²/λ is used so often at FR2 that it is sometimes copied onto a range sketch without the phase error it already accepts, or without checking whether D is the largest dimension that is actually radiating. At 28 GHz the wavelength is c/f = 10.707 mm. The range length then scales with the square of the antenna size:
- 10 cm AUT: 2D²/λ = 1.87 m
- 15 cm AUT: 4.20 m
- 20 cm AUT: 7.47 m
- 30 cm AUT: 16.81 m
For comparison, the same 20 cm aperture at 10 GHz (λ = 29.98 mm) only needs 2.67 m. A lab that was comfortable at X-band can run out of room as soon as the same horn moves to 28 GHz, and a compact-range or a near-field scanner is then a geometry decision, not a luxury.
What 2D²/λ actually buys. The criterion corresponds to a maximum path-length difference of about λ/16 from the centre of a uniformly illuminated aperture to its edge, which is a 22.5° phase taper over the aperture. It is a commonly accepted starting point, not a zero-error condition. If a pattern or a gain figure is being signed to a tighter phase budget — some labs work to λ/32, which doubles the range — 2D²/λ is already too short. We would write the rule that was used (2D²/λ, 2 × 2D²/λ, or a stated phase error) on the same line as the gain or the sidelobe number, because a 0.3–0.5 dB disagreement between two ranges is often this choice, not the AUT.
What "D" should be. D is the largest radiating dimension, not the flange and not the absorbing box. A 15 cm horn with a 25 cm fixture plate still has D ≈ 15 cm if the plate is not radiating; a 15 cm patch array on a 30 cm ground plane may have to be treated as the larger of the array and the currents that actually run on the ground. If you are unsure, use the larger number and say so. Using the patch width alone, then being surprised that the measured phase across a probe scan is worse than 22.5°, is a definition problem.
Two checks that are cheaper than arguing about the formula:
1. Amplitude taper and quiet-zone size. Even a range that meets 2D²/λ can still put a 1 dB amplitude taper across the AUT, which will move a gain or a beamwidth number by a few tenths of a dB. Measure the quiet zone with a probe or a small horn at the AUT position, and write the taper and the ripple next to the far-field claim.
2. A known-gain standard. A substitution horn with a calibration that covers 28 GHz, measured in the same range length, is the acceptance we would sign before a DUT gain is written down. If the standard already disagrees with its calibration by more than the stated range uncertainty, lengthening the range or moving to a compact range is the next step, not a software correction on the DUT.
A compact range or a near-field scan is the practical answer once 2D²/λ no longer fits the room. It does not remove the need to write down the effective far-field criterion, the quiet-zone ripple, and the standard-gain check. Those three lines are what make a 28 GHz gain figure comparable between two labs.
Open question: at 28 GHz, do you still use 2D²/λ as written, do you double it, or have you already moved that size of AUT to a compact range or a near-field scanner — and how large a gain disagreement have you seen between a "just 2D²/λ" range and a longer one?
— rfforum 编辑部 / Editorial Team