Measuring coupler directivity with a 20 dB load: the number you wrote down may be the termination, not the coupler

Started by LunaWei · 2026-11-27 02:10 UTC · 7 views · 1 replies
L Admin
[Q&A / FAQ, rfforum Editorial Team]

A directional coupler's directivity is the reason a reflected-power or a return-loss measurement is trustworthy. It is also one of the easier figures to measure wrongly, because the termination you use to "kill" the forward path is itself a reflector, and a 20 dB load will hide any coupler that is better than about 20–25 dB.

Recall the definitions, so the arithmetic is in the open. Coupling is the ratio of the coupled-arm power to the input power, with the output port well terminated. Isolation is the ratio of the isolated-arm power to the input power, with the output well terminated. Directivity is isolation minus coupling, in decibels: a 20 dB coupler with 45 dB of isolation has 25 dB of directivity. Residual error when you use that coupler to measure a high return loss is set by the directivity: a 25 dB directivity and a 30 dB DUT return loss are already comparable, and the reading can be almost anything in between, depending on the phase of the residual.

The trap is the load used to measure isolation (and therefore directivity). Suppose the coupler under test has a true directivity of 35 dB, and you terminate the output with a load whose own return loss is 20 dB. A fraction of the forward power reflects from that load, returns to the coupling mechanism, and appears on the isolated arm as if it were leakage. The apparent isolation, and therefore the apparent directivity, cannot look much better than the load. In a well-known first-order picture, a 20 dB termination puts a floor under the measurement at about that same 20 dB of equivalent directivity, give or take the coupling factor and the phase, and a 35 dB coupler will then be reported as "about 20 dB." The coupler was not out of specification. The load was the instrument.

What we would do instead:

1. Use a termination whose return loss is well above the directivity you hope to see, at the frequency of the test — a 40 dB sliding load, a precision 3.5 mm or 2.92 mm termination with a calibration, or a sliding load whose ripple you can separate from a fixed residual. A general-purpose 20 dB, 1 W load that was bought for a power amplifier is not this part.
2. If only a modest load is available, move it (a short airline or a sliding load) and watch the isolated-arm reading ripple. The ripple amplitude is set by the load; the average, or the residual after the ripple is taken off, is closer to the coupler. A single reading with a fixed 20 dB load is not a directivity number.
3. Write the load's return loss, the frequency, and whether a sliding-load subtraction was done, on the same line as the directivity. A later engineer can then tell a 25 dB coupler from a 25 dB load.

The same caution applies in the opposite direction: if you use the coupler to measure a DUT return loss, the DUT number cannot be trusted much beyond the coupler's true directivity. A 20 dB-directivity coupler will not settle an argument about whether a filter is 25 dB or 30 dB return loss.

Open question: what load, or what sliding-load method, do you actually require before you will sign a coupler directivity above 25 dB, and have you ever had to withdraw a "failed directivity" once a better termination was used?

— rfforum 编辑部 / Editorial Team
L Admin
Editorial addendum — a 20 dB load cannot sign a 35 dB directivity.

If the termination used to measure isolation is itself a 20 dB reflector, the apparent directivity will sit near that floor, and a good coupler will look bad. Do you already require a sliding load, or a termination whose own return loss is well above the directivity you hope to see, before signing the figure?

— rfforum 编辑部 / Editorial Team