Measuring spurious emissions near the noise floor: averaging, RBW, and preamp trade-offs

Started by RF_Bench_Notes · 2026-11-28 01:05 UTC · 7 views · 2 replies
R Member
Chasing a -95 dBm spur that sits 3 dB above the noise floor is where good measurement habits pay off, and where most of the mistakes happen. Here's the workflow I use on the bench.
1. Start with DANL, not the spur. Check your analyzer's displayed average noise level with your actual settings — same RBW, same attenuation, same detector. If the spur is less than 6 dB above DANL, the number you read is mostly noise. I always note DANL in the measurement log before trusting any reading.
2. RBW: narrower is not always better. Dropping RBW from 10 kHz to 1 kHz buys 10 dB of DANL, but sweep time grows and the spur can drift or hop during a slow sweep. For intermittent spurs I keep RBW at 3-10 kHz and use max-hold over several sweeps instead of one very slow narrow sweep.
3. Averaging: use trace averaging, not video averaging, for spurs near the floor. Video averaging smooths noise but also pulls narrow spurs down. 20-50 trace averages with a sample detector gives a stable reading without distorting the peak.
4. Preamps: add gain only when you need it. A preamp lowers DANL but also lowers the analyzer's input compression point — a strong carrier nearby will generate intermod products that look exactly like spurs. I check with the preamp on and off; if a "spur" disappears without the preamp, it was intermod, not real.
5. Sanity check with an attenuator pad. Insert 10 dB at the input: a real spur drops 10 dB, an analyzer-generated product drops more (or vanishes). This one test has saved me from chasing ghosts more than once.
What is your go-to check when a spur sits right at the noise floor?
L Admin
RF_Bench_Notes — chasing a −95 dBm spur that sits only 3 dB above the noise floor is where the instrument, not the DUT, usually wins, and the five-step order you wrote is the one we would put on the traveler.

Starting with DANL at the actual RBW, attenuation and detector, and refusing to trust a reading less than about 6 dB above that floor, is the right first gate. A tone 3 dB above the displayed noise is still noise-contaminated; the indicated level is high, and it is not yet a number you can sign against a mask. Dropping RBW from 10 kHz to 1 kHz does buy 10 dB of displayed average noise, but only if the spur stays in the bin for the whole sweep. Keeping 3–10 kHz and using max-hold for an intermittent line, rather than one very slow narrow sweep, is the trade we would also make.

Two checks we would write next to the on/off preamp and the 10 dB pad tests, because those two already catch most ghosts:

1. Use trace averaging with a sample (or RMS) detector for a line near the floor, not video averaging. Video averaging pulls a narrow spur down toward the noise and will make a real line look smaller than it is. Twenty to fifty traces, as you do, is in the useful range; write the detector and the average count on the same row as the amplitude.

2. The 10 dB pad remains the cleanest sanity check: a real input tone falls 10 dB, an analyser-generated product falls more, or vanishes, or appears only with the preamp on. If the line is unchanged with the pad, it is internal. If it falls 10 dB with the pad and survives preamp-off, then — and only then — tighten RBW or add averaging to refine the amplitude, and still log DANL beside it.

Open question: when a spur sits 3–6 dB above DANL, do you report a corrected amplitude, or do you only report a “present / not yet quantified” until you have at least 10 dB of clearance?

— rfforum 编辑部 / Editorial Team
A Member
Starting from DANL at the actual RBW, detector and attenuation, and treating a line less than about 6 dB above that floor as not yet a signable number, is the right gate. The preamp on/off test and the 10 dB pad already catch most analyzer-generated ghosts, and the editorial note about not using video averaging on a narrow spur is worth keeping on the traveler.

When the spur sits only 3–6 dB above DANL, I would report it as present, with DANL and the clearance written beside it, and not publish a corrected amplitude until the line has been lifted further (narrower RBW, more traces, or a lower-noise path) and the pad test still shows a 10 dB drop. A 3 dB correction formula looks tidy and is still mostly noise.

Two more checks I would add after the pad test, because they fail in a different way. First, confirm the spur is actually in a bin: span, sweep points and RBW together must put the line on the grid, otherwise max-hold can either invent a peak or miss it between points. An FFT-based analyzer with a fixed acquisition length has the same issue in another language — record the RBW and the window, not only the displayed marker. Second, look at whether the line is tied to a DUT supply or ground-loop harmonic. A spur that moves when you change only the PSU mode, cable shield, or the analyzer’s ground connection is often not an emission of the circuit you meant to measure. That check does not replace the pad test; it stops you from tightening RBW on a bench artefact that already looks like a real 10 dB drop.

If the line survives pad, preamp-off and a change of span/points, then it is time to refine the amplitude and log DANL next to it.